Proposed changes to ANSI/HPS N13.32 "Performance Testing of ExtremityDosimeters"
In order to provide the health physics community information
regarding the development of standards, and, to obtain pertinent
feedback that can be incorporated into the standards development
process, various venues are being pursued. Below are draft positions
taken by the N13.32 Working Group. Formal presentations are being
made, including the recent 22nd Annual International Dosimetry
Symposium and National Dosimetry Records Conference, and the upcoming
Thermo Electron Users Meeting, and, the FL HPS and AAPM Meeting.
The N13.32 Working Group had identified twelve major issues for
consideration. The following paragraphs describe how the Group
implemented the draft resolution of the issues. Some of the issues
are treated in greater detail in the appendices that were written to
provide greater insight and convenience. The working group made the
most significant changes to the 1995 version of the standard in the
areas of test categories and test criteria.
The working group attempted to harmonize the test categories with
those in the sister standard, ANSI HPS N13.11-2001. Particularly,
the photon test categories in the protection level dose range were
combined so that the previous test categories for low-energy and high-
energy photons, categories II and III, are now both included in test
category II for photons. In addition, the number of x-ray fields
available for testing in the photon category was increased from four
x-ray fields and one high-energy photon field to six x-ray fields and
two high-energy photon fields. The beta category now included as
category III, remains unchanged except for the addition of a low-
energy beta source as an alternative to 204Tl.
The working group considered the inclusion of a neutron-testing
category based on the recommendation in the journal of the ICRU,
Volume 1, No. 3 (2001) "Dosimetry for Irradiation of the
Extremities." At this time, though, the working group felt that the
theoretical basis of neutron dosimetry to extremities has not reached
a level of national and international agreement to promote the
practice of neutron extremity dosimetry by inclusion of a testing
category.
At the request of the dosimetry community, two additional test
categories were added, photon-mixture and beta/photon-mixture
categories. These categories were added to accommodate test
participants submitting wrist dosimeters with the capabilities to
interpret Hp (0.07) in mixed fields. To clarify the intent of the
new categories, the working group also added an option, requested by
the test participant, for blind testing. Participation in the
mixture categories is only available to participants who submit
dosimeters for blind testing. Normal testing, as in the last version
of the standard, includes only categories I through III with the
testing source identified to the participant before analysis. The
working group does not recommend blind testing for single element
extremity dosimeters.
The range of testing doses has remained unchanged, although the
working group agreed to adopt personal dose equivalent at 0.07 mm as
the basis of the test of performance. Guidance provided in
international documents indicates that directional dose equivalent is
a suitable approximation to personal dose equivalent. Conversion
coefficients for photons, listed in ICRU 67, were used with digitized
spectra for the NIST x-ray beams to determine coefficients to convert
air kerma to personal dose equivalent for the x-ray testing fields.
Research has shown that the dose rate at 0.07 mm used for beta
particles incident on the slab phantom is applicable for use with the
rod and pillar phantoms. In selecting personal dose equivalent at
0.07 mm, the working group chose to exclude a discussion of lens-of-
the-eye (LOE) dose. The group concluded that it was inappropriate to
include LOE dose as part of a standard addressing extremity dose.
For practical purposes, the polylmethyl methacrylate (PMMA) rod
phantom will continue to be used for testing of finger dosimeters.
The pillar phantom with the aluminum insert will be replaced by a
PMMA phantom of the same dimensions. The backscatter from the PMMA
pillar phantom in photon fields is more similar to the ICRU tissue
phantom for which the conversion coefficients have been determined.
The working group considered the use of the water-filled PMMA pillar
phantom by conducting a controlled irradiation of extremity
dosimeters on PMMA, water-filled PMMA, aluminum-core PMMA and
Styrofoam pillar phantoms. The results did not yield evidence that
the backscatter conditions to the dosimeter were significantly
enhanced using the water-filled PMMA phantom over the PMMA phantom.
The criteria for testing the performance of personal extremity
dosimeter systems has historically taken a systematic approach, i.e.,
testing the performance of a group of dosimeters rather than basing
the test on individual dosimeter results. This philosophy was
continued for this version of the standard; however, the model for
testing was modified. In the past, the test of performance was based
on a) the sum of the absolute value of the bias and standard
deviation of fifteen dosimeters irradiated in a single test category
and b) limits on the individual values of bias and standard
deviation. In this standard, an approach was taken to combine the
two values in quadrature consistent with current theory in quality.
The approach has decreased the area of acceptable performance by less
than three percent.
The working group modified the required ancillary tests to further
distinguish between type tests and periodic tests. The requirements
for the lower limit of detection and angular response testing were
removed from this standard, as they constitute one-time tests that
should be performed upon the initial implementation or modification
of a dosimeter system. Recommended protocols for those studies are
described in the attached appendices. In addition to those studies,
the working group modified the standard to also recommend the study
of uncertainty for each dosimeter system. Guidance, based on the ISO
Guide to Uncertainty in Measurements, is given in the appendices for
the approach to uncertainty analysis.
The Working Group anticipates ballotting this standard sometime
during the next 6 months. Additional updates will be provided to the
industry for feedback.
If you have specific comments, please provide them to Rick Cummings
CUMMFM@inel.gov .. Chair, N13.32 Working Group, or directly to me.
Sandy Perle
Chair, HPSSC
************************************************************************
You are currently subscribed to the Radsafe mailing list. To unsubscribe,
send an e-mail to Majordomo@list.vanderbilt.edu Put the text "unsubscribe
radsafe" (no quote marks) in the body of the e-mail, with no subject line.