1.  The mock iodine source issue (particularly the Ba-133/Cs-137 source) was
a particular problem for us at Oak Knoll and Bethesda.  The problem was the
differential ratio of activity due to the relatively short half-life of
Ba-133 compared to Cs-137.  True, it's still in the "years" range, but you'd
be surprised how many over 10 year old "mock" iodine sources are out there!

2.  I subscribe to the at least once a year calibration using the real stuff
approach (I-131/I-125).  The summing issue for the well crystals commonly
used in clinics today is not trivial.  Also, a standard longer lived source
(such as the Cs-137 source we used for daily dose calibrator checks), prefer-
ably reasonably well matched for the photon energy (I-129 isn't too bad for
I-125), is reasonable assurance against missing significant window drifting.
Significant being the level specified in your license for that assay system.
(Of course, just to be sure when dosing, we ordered enough to check the
window everytime we scheduled bioassay--the real stuff that is.)

----------------------------------------------------------------------
The opions expressed above are    |  The opinions above also do not
those of the author alone and do  |  represent those of US Department
not represent those of the        |  of Defense, Defense Nuclear
Stanford University or the US     |  Agency, US Navy, Metropolitan
Department of Energy.             |  Edison, Porter Consultants, etc.
----------------------------------------------------------------------



                         - - - - Forwarded Text - - - -


Received: from VMD.CSO.UIUC.EDU by SLACVM.SLAC.STANFORD.EDU (Mailer R2.08
 R208004) with BSMTP id 1756; Thu, 28 Jul 94 14:24:28 PST
Received: from UIUCVMD (NJE origin SMTP@UIUCVMD) by VMD.CSO.UIUC.EDU (LMail
 V1.2a/1.8a) with BSMTP id 8520; Thu, 28 Jul 1994 16:25:21 -0500
Received: from vixen.cso.uiuc.edu by vmd.cso.uiuc.edu (IBM VM SMTP V2R2)
   with TCP; Thu, 28 Jul 94 16:25:20 CDT
Received: from romulus.ehs.uiuc.edu by vixen.cso.uiuc.edu with SMTP id AA01049
  (5.67b/IDA-1.5); Thu, 28 Jul 1994 16:24:32 -0500
Received: by romulus.ehs.uiuc.edu (NX5.67d/NeXT-2.0)
        id AA20212; Thu, 28 Jul 94 16:23:40 -0500
Date: Thu, 28 Jul 94 16:23:40 -0500
Message-Id: <9407282123.AA20212@romulus.ehs.uiuc.edu>
Errors-To: mandel@romulus.ehs.uiuc.edu
Reply-To: radsafe@romulus.ehs.uiuc.edu
Originator: radsafe@romulus.ehs.uiuc.edu
Sender: radsafe@romulus.ehs.uiuc.edu
Precedence: bulk
From: FRAMEP@orau.gov
To: Multiple recipients of list <radsafe@romulus.ehs.uiuc.edu>
Subject: Re[2]: I-125, 131 sources
X-Listserver-Version: 6.0 -- UNIX ListServer by Anastasios Kotsikonas
X-Comment:  Radiation Safety Distribution List



          It is acutely embarassing that I can't provide more info
          about mock iodine than what follows. The stuff was first
          developed by the late great Marshall Brucer when he worked
          at ORAU in the 1950s. The historical collection here
          contains one of his primary mock iodine standard sets
          assayed in 1959.  As Albert surmized, the additional
          component is Cs-137.  The set says that the ratio of Ba-133
          to Cs-137 was 12 to 1.  An Amersham catalog I have lists
          "simulated iodine" and says it contains Cs and Ba but
          doesn't give the ratios.  I'm sure they can provide the
          data.  Interestingly, the catalog warns that the mock source
          emits a bunch of low energy photons that could lead to
          summation losses  - something that could be especially
          significant with a well counter.  They suggest a cadmium
          filter be used when counting the mock source to cut out its
          low energy component.
             ANSI 42.13 might say something about this but I don't
          have a copy and can't be sure.

          Salutations and regards

          Paul

          HP trivia: Otto Hahn died on this date in 1968
          and last Tuesday, the 26th, was the 99th wedding aniversary