[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: roentgen to rad factor
Roentgen is a unit for radiation intensity.
rad is a unit for deposited dose and depends on the absorber.
You can calculate deposited dose (in rad dor other units) for any intensity
(in roentgen or other units) with the PC prgram PHOTCOEF. Look it up at
www.photcoef.com
Paul Kehler
photcoef@aol.com
At 05:40 PM 11/14/97 -0600, you wrote:
>>X-Sender: viru@therad
>>Date: Fri, 14 Nov 1997 14:55:53 -0600
>>Reply-To: Medical Physics Mailing List <MEDPHYS@LISTS.WAYNE.EDU>
>>Sender: Medical Physics Listserver <medphys@lists.wayne.edu>
>>From: Ramasamy Virudachalam <viru@THERAD.RPSLMC.EDU>
>>Subject: Re: roentgen to rad factor
>>To: Multiple recipients of list MEDPHYS <MEDPHYS@LISTS.WAYNE.EDU>
>>
>>
>> ** Mail from Medphys Listserver **
>>If you reply to this message, it will be posted on Medphys for all the
>>subscribers to review ...
>>
>>
>>>As an aside (and given that I currently practice diagnostic physics), I
>>notice in Khans' _The Physics of Radiation Therapy_ that this factor is
>>.876 rad/R while Johns and Cunninghams gives .873 rad/R. Which witch is
>>which?
>>>
>>Bill,
>>The correct factor is 0.876 rad/R.
>>The difference you observed between the values reported in Kahns and Johns
>>books stems from a change in W/e from 33.85 eV/ion pair to 33.97 eV/ion
>>pair. For details refer to Attix's book on Introduction to Radiological
>>physics and Radiation Dosimetry, p72.
>>
>>
>>R.G. Virudachalam, Ph.D.
>>Rush Pres. St. Luke's Med. Ctr., Chicago
>>312-942-5751
>>
>
>
>