[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re[4]: Lead Aprons in Dentistry
>This is per a posting from a member on the MedPhysics LiustServer, a
>state inspector.. He states that the older houysings had more
>shielding comparec to the newer ones. Here is what he posted..
>better using his words than mine.
>>Housing leakage is definitely measurable from dental tube heads.
>>The FDA says that leakage radiation shall not exceed 100 millirem
>>in one hour at a distance of 1 meter.
Actually - more precisely stated - it is 100 mR in any one hour at 1 meter
AT MAXIMUM CONTINUOUS KV AND MA! It would be meaningless to state the
limit in terms of mR/hr (which it seems the person originating the message
would prefer!)
Certainly this DOES mean that the manufacturers COULD allow leakage rates
as calculated - but they don't!
I have had lots of experience measuring leakage from dental x-ray units -
but certainly not as much as a state inspector might. AND when I do such
a measurement I remove the collimator and put a lead blocker on the port so
as to avoid measuring scatter. In ALL the measurments I have ever done I
have NEVER seen an instantaneous leakage rate of anywhere near 100 mR/hr at
1 M at OPERATIONAL kV and mA - let alone at continuous (which it is
IMPOSSIBLE to do from accessible controls). Even doing measurements with
scatter off a phantom, at 1 M - I seldom see more than a couple 100 mR/hr -
instantaneous rates.
Also - 0.5 second would a relatively long exposure for modern techniques.
More like 0.1!
A dental unit would most likely melt into a puddle before there could be
any sort of exposure of concern from leakage.
Oh yeh - and the distance to the chest is irrelevant - since the apron
required is a lap apron and not the normal apron worn in x-ray labs.
Certainly a normal apron could be used - but it's not required.
>>The use of "in one hour"
>>instead of "per hour" allows the manufacturer to stipulate a duty
>>cycle for the machine. One duty cycle I've seen is that the
>>machine should only be operated for one second out of every 60.
>>This would allow a instantaneous rate of 6000 mR/hr , or 1.7 mR/sec
>>at 1 meter. Typical exposure times depending on mA and cone length
>>are on the order of 0.5 sec and if one assumes a distance of 0.25
>>meters to the patients chest and it looks like you could get 3.4 mR
>>per shot. 14 films or so in a full mouth series yields 47 mR at
>>0.25 meters.
Ted de Castro
tdc@ehssun.lbl.gov
University of California Lawrence Berkeley Laboratory
Bldg 75B Rm 112A
Berkeley, CA 94720
(510) 486-5256
(510) 486-5506 - FAX