[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Equation for MDA
We have a digital radiation monitoring system installed. A question has
arisen as to what is the most correct method of determining the Minimum
Detectable Activity ( MDA ).
The vendor manual list the MDA = (a * ( B/(2*T) )^0.5 ) * C
a = confidence intervals
T = instrument response time - because this is a digital instrument,
this varies with the detector count rate
T = 10 if B < 100 cpm
T = 1000/B for 100 < B < 100,000 cpm
T = 0.01 * B B > 100,000 cpm
B = Background Count Rate
C = uCi/cc per cpm conversion factor
This appears to be based on the treatment of the system as a Count Rate
Meter ( digital Frisker ! ) and determining the standard deviation as you
would for that type of instrument. This appears to be an Lc value converted
to Activity.
Currie (NUREG/CR-4007) on page 91, appears to indicate a different method of
calculating MDA that is essentially :
MDA = ( ( 2.71 + 3.29 * (Sb)^0.5) * C
Sb = Standard deviation of the background
C = uCi/cc per cpm conversion factor
this assumes that the Background and sample counting times are the same.
The questions are :
1 - What is the correct method for determining the MDA for this type of
instrument.
2 - Which value should be used for setting the "alarm setpoint".
3 - does anyone know of a reference that describes the derivation of the
first equation
Thank you all for your time and consideration.
Ron LaVera
RLaVera@entergy.com
914-736-8433
914-736-8419 FAX
Entergy Nuclear Northeast
Entergy Nuclear Operations, Inc.
Indian Point 3 NPP
P.O. Box 308
Buchanan, NY 10511
************************************************************************
You are currently subscribed to the Radsafe mailing list. To unsubscribe,
send an e-mail to Majordomo@list.vanderbilt.edu Put the text "unsubscribe
radsafe" (no quote marks) in the body of the e-mail, with no subject line.
------------------------------