Sorry if I jump in a bit late, but my grain of salt:
not the detector, the electronics
proportional counters use <very> high input impedance opamps, to measure
very low currents -
typically the 8500A (Datel-Intersil, now Dawn) CMOS chip, bad in all specs
except some quite spectacular 10fA bias at 25 C
good manufacturers even screen these units to 3 fA
anyhow, the only thing wrong is that in MOS (and JFET) technology bias
increases <dramatically> with temperature, if I remember right one factor
of ten every three degrees Celsius - really - three log units per 10
degrees
some people cool their chips with Peltiers, but I fear not in this case
so I would suspect the chip bias rather than tempco in the HV, which should
be well and easily compensated in a good design
why does the current stay high on cooling? thermal inertia? self heating?
in aging machines stray currents due to condensation on the PCB (of water
and alcohols) aided by poor layout have been known to knock out the specs
upon cooling:
...sticking a silica gel pouch in the case probably would not hurt...
(pH meters I was making in Spore and Mauritius would not work at all in
Rhode Island until we let them dry a couple days, they arrived all wet
inside)
simple morale: happens all the time: people trust machines too much
we build physical theories to explain strange observations when often these
are just instrumental artifacts
guys get fried because their GMs all saturate and read zero
etc.
Marco Caceci
http://radal.com
> -----Mensaje original-----
> De: Cehn@AOL.COM [SMTP:Cehn@AOL.COM]
> Enviado el: viernes, 24 de agosto de 2001 20:26
> Para: radsafe@list.vanderbilt.edu
> Asunto: Floor monitor - background increasing
>
> Thanks for your reponses to my post (see below). The cause of the
increasing
> background appears to be a temperature effect on the detector. Ludlum
admits
> that for these gas filled proportionals, efficiency goes up with
temperature.
>
>
> I would have guessed it would go down, based on the decreasing density of
the
> counting gas with increasing temperature. But other factors (not yet
> revealed) are at work.
>
> The lesson is choose your high voltage carefully. If you're on a flat
> plateau, temperature effects should be minimal.
>
> Oringinal post:
> "I have a Ludlum 43-37 floor monitor surveying a large concrete pad -
> outdoors. On very hot days (>105), the instrument background goes up.
We
> take it into a cool trailer, cool it down, and background stays up. Any
> ideas?
>
> My favorite theory so far is radon coming off the concrete in the heat,
and
> the daughters sticking to the detector. That would also account for the
> backgrounds returning to normal the next morning."
>
> Joel I. Cehn, CHP
> cehn@aol.com
> Oakland, CA
> <<Archivo: ATT00009.htm>>
************************************************************************
You are currently subscribed to the Radsafe mailing list. To unsubscribe,
send an e-mail to Majordomo@list.vanderbilt.edu Put the text "unsubscribe
radsafe" (no quote marks) in the body of the e-mail, with no subject line.