To those interested in misadministrations:
Initially the allowed variation was 10%.
At one time the NRC staff proposed to stop calling them misadministrations
and the Commissioners indicated that they felt a spade should be called a
spade.
In 1994 the allowed variation was increased to 20% of the prescribed dose
but between 10 and 20% were called "recordable events". They must be
documented and investigated but are not reported to NRC at the time. They
are reviewed by inspectors when they visit.
In the revised part 35 tentatively scheduled to be published in the Federal
Register in October the renaming is accomplished they will now be known as
"medical events" a medical event will be reportable. Generally 20% is still
the standard but also 5 rem for whole body and 50 rem for organ or skin dose
is needed to make it a "medical event".
The "recordable event" is dropped in the revised regulation.
Editorial comment- Though I generally agree with the relaxation of many of
the prescriptive medical regulatory requirements I think recording events
over a certain threshold that are not medically significant but the result
of mistakes should be done, not for regulatory review but for quality
assurance reviews by the licensee. I have not heard that NRC inspectors
have made a lot of citations out of recordable events so I guess I would be
in favor of leaving it in the regulation.
The Society of Nuclear Medicine and the American College of Nuclear
Physicians have been fighting adoption of the revised regulation because
they feel that diagnostic nuclear medicine is over regulated and the
revision does not substantially change that. They managed to get the Senate
appropriations bill for the NRC to include language to prohibit use of funds
to enforce the revised part 35 if it is enacted. The House version does not
include the language so a conference committee will resolve that. The Bush
administration has stated it's position that the new regulation should be
enacted as it provides some relief from the current part 35.
More editorial comment- I do not have figures but I think Nuclear Medicine
does fairly well in significant misadministrations, meaning ones that cause
physical harm or significant risk of harm.
Another difference between power plant health physics and NM HP is that
radioactive materials and radiation is deliberately applied to humans in NM
which is not the case in power plants. When there is a monumental mistake,
or rather series of mistakes, you get a Three Mile Island. In NM when there
is a big enough mistake in therapeutic use you get a death. These have been
relatively rare.
IS MORE REGULATION NEEDED? In 1994 the FDA issued an advisory regarding
"occasional but at times sever radiation induced burns..." from
fluoroscopically guided procedures. These procedures are generally
interventional procedures that require placement of a stent or other
invasive life saving procedures. Without regulatory intervention that
advisory, a later follow up and the concern of the accreditation agency, the
Joint Commission on Accreditation of Health Care Organizations, fondly known
as JCAHO or Jake [8-}> means that many facilities now require training for
non-radiology physicians directing fluoroscopy procedures, recording of beam
times and dose estimates. Note that at our facility most of the non
radiology physicians doing fluoro are doing procedures well under 30 minutes
of beam time. Generally, there will be little risk of acute skin effects
below 2 hours of beam time. There are a host of variables, not the least of
which is how much the area in the beam changes during the procedure. As far
as I know there are no regulations requiring us to do anything about the
potential for fluoroscopy skin damage.
Any opinions expressed are mine alone and do not necessarily represent
those of the Denver VA Medical Center, The Department of Veterans
Affairs, or the U.S. Government.
Peter G. Vernig
Radiation Safety Officer, VA Medical Center, 1055 Clermont St. Denver,
CO 80220, ATTN; RSO MS 115
303-399-8020 ext. 2447, peter.vernig@med.va.gov [alternate for business
- vernig.peter@forum.va.gov; private - peter_vernig@hotmail.com] Fax
303-393-5026 [8 - 4:30 MT service] Alternate Fax 303-377-5686
"...whatever is true, whatever is noble, whatever is right, whatever is
pure, whatever is lovely, whatever is admirable, if anything is found to
be excellent or praiseworthy, let your mind dwell on these things."
Paul
-----Original Message-----
From: Perrero, Daren [mailto:Perrero@IDNS.STATE.IL.US]
Sent: Thursday, September 06, 2001 12:29 PM
To: radsafe@list.vanderbilt.edu
Subject: RE: medical misadventures
Bill
The other side of this argument is that a fair number of human errors that
end up as medical misadminstrations are violations of regulations that have
no place being burdened upon licensees in the first place. No "major event"
has even occurred. The "errors", in fact, have no impact on the health of
the patient involved. The range of effectiveness for most therapeutic doses
are enormous and whether you are off by 20% or more is irrelevant to the
outcome, or better yet the administration of radiation is a means of
providing "additional assurance" that the main treatment method (i.e.,
surgery) is successful, yet by regulation the "error" is a reportable
misadministration. No harm, no foul, I say!
The rad protection program at the facility is, more often than not, just
fine and the misadministration is not a valid indication of the quality of
work being done by the Med Physicist/RSO at these facilities. Med HP's
spend their entire day trying to improve "human performance", but like
everyone else they have to allocate limited resources. I would rather see
them spending their time with the CT Scanner, Mammography Unit and LINAC and
other X-ray sources since there are far more medical procedures conducted
with these forms of radiation than in the busiest Nuc Med Dept.
If its really about trust, and not dose, then I suggest the medical folk
stop being required to wave red flags when the patient impact is negligible
in the first place.
************************************************************************