Re: Removable Contamination Surveys



Title: Removable Contamination Surveys
Tom and other radsafers
 
I'm not an instrument person but I seem to recall that 100 cm^2 was the "best" beta counting area for the instruments at one time, and this was translated into regulations. The 100 cm^2 has apparently been extraplolated to the "smearing" area.
 
Experts like Jack Fix and Joe McDonald at PNNL could likely identify the correct background.
 
John R Johnson
Vancouver B C.
----- Original Message -----
Sent: Monday, February 04, 2002 11:03 AM
Subject: RE: Removable Contamination Surveys

One comment...note that (whether good or bad) in the medical world, 10 CFR Part 35.70(h), 35.315(a)(7) (medical use licensees) requires contamination survey results to be expressed in "dpm per 100 cm^2" (this implies that a 100 cm^2 area be smeared) when working with byproduct radiopharmaceuticals.
 
Tom Huston, PhD, CHP
-----Original Message-----
From: Daniel Mackney [mailto:photond@EARTHLINK.NET]
Sent: Monday, February 04, 2002 11:06 AM
To: Jenkins, Peter; radsafe@list.vanderbilt.edu
Subject: Re: Removable Contamination Surveys

We were taught in the US Navy nuke program that the regulation was alpha or beta-gamma contamination/1cm^2. And if you look in the regs that's what it says.
However, the 100 cm^2 became a convention based on the princple background of K40 is around 5  to 50 uuCi/100 cm^2 . A count of 450 uuCi/100cm^2 was 10 to 100 times normal background or 100 cpm above background, with a DT-304 frisker probe at 1/2 inch above the surface, which is where we were told to set our GM detectors to alarm at. This allowed for random fluctuation in background without continuous alarms. This comes from NAVSEA 0989-015 printed in 1975 Rev 4 and first appeared in 1957 in the Navy's RadCon Manual NAVSHIPS 0153.
The motion of the smear as stated by others is based on two fingers pressed together on a smear makes a 1" wide mark. Now if that 1" is dragged over 16" then approx. 100cm^2  has been covered. The 'S' shape was to ensure the area was "averaged", which is not done is a straight line is done.
 
If you look at 49CFR 173.443 a wipe of up to 300cm^2  is permissible.
 
Dan Mackney
Dirrector of Radiochemistry
Wate Stream Technology
----- Original Message -----
Sent: Friday, February 01, 2002 2:50 PM
Subject: Removable Contamination Surveys

Can anyone give me some insight (and possible source documents) into the reasons for the 100cm^2 standard (or suggestion?) used for removable contamination surveys?  Why is it 100cm^2 and not 1,000cm^2, or something smaller or larger?  Is there anything wrong with increasing the swiped area to increasing the likelihood of detection? 

Specifically, assume a removable contamination limit is set at 200 dpm/100cm^2 for a laboratory and a researcher wishes to use a survey meter to count the swipes.  According to the manufacturer's formula and values (for efficiency of the radionuclide, etc), the meter has an MDA of 400 dpm.  Is there anything wrong with performing the survey over 200 cm^2 so that the detection limit would meet the removable contamination limit?

Thanks,

Pete Jenkins

Confidentiality Note: The preceding e-mail message (including any attachments) contains information that may be confidential, protected by applicable legal privileges, or constitute non-public information.  It is intended to be conveyed only to the designated recipient(s).  If you are not an intended recipient of this message, please notify the sender by replying to this message and then delete it from your system.  Use, dissemination, distribution or reproduction of this message by unintended recipients is not authorized and may be unlawful.