Re: start up calibration facility
>I'm looking for recommendations and information on starting up
>a calibration facility for portable detectors here at the
>University of Maryland. The RSO here about 10 years ago used to
>calibrate their own detectors using a Cs-137 source with an attached
>sliding Attenuator from J L SHEPHERD, out of Glendale, CA.
>I have seen smaller versions for around 3500 dollars, and we
>have a pulser for cpm meters; however I have no experience in
>setting up or calibrating other than what I have read in a text.
Steve,
If I were in your shoes?
Keep the Shepherd source. It's going to remain useful for many years to
come. Make sure the distance range is updated for decay periodically, and
that it's long enough for near-background ranges of instruments. Our lab is
a bit cramped; we have a Bicron Micro-Rem(TM) meter, and we just have to
guess for the lowest scale, which goes from "0" to 20 microrem/hr, because
our range isn't long enough. (It's pretty narrow, too; if they're assigning
you new space for this start-up facility, be sure to ask for a range wide
enough to minimize scattering from the walls) The new 100 mrem/year dose
limit makes low-range accuracy important.
Also be aware of regulatory requirements: The results of all RADIATION
measurements are to be recorded in DOSE or DOSE EQUIVALENT units. So when
setting up the range, make sure to use the *rem* gamma constant, not the
*Roentgen* gamma constant we all know and love. The results of all
CONTAMINATION measurements have to be in ACTIVITY units (not CPM and not
mrem/hr) like microcuries or DPM. Since most of our instrument users frisk
their labs, and themselves, for contamination, and your university is likely
to be similar, I think this kind of calibration is more important than using
the Cs-137 range: Make sure you have a good supply of point-like sources
and wide-area sources of long-lived analogs of the nuclides used by the
instrument's main users. Use these sources to give the users conversion
factors from meter response (cpm, cps, etc.) to DPM and/or DPM/100cmcm for
the nuclides they use. A standard probe-surface distance, for this
efficiency, is 1 cm.
By all means get a copy of NCRP-112, too. We're still waiting for ours.>:-|
Albert Lee Vest avest@magnus.acs.ohio-state.edu
health physicist Office of Radiation Safety
(614)292-0122 The Ohio State University