Re: RE: Keithley chambers for equip. performance evaluations of Panorex units
>> Does Keithley make a CT pencil type detector with a long cable? I
>>remember
>> using a RADCAL 1015 with a CT probe taped vertically in front of the
>> film cassette plate, with enough play in the cable to prevent the
>> probe from being torn loose during rotation.
I wouldn't tape that probe to the film plate for 3 reasons:
1: I THINK it is 1 cm in dia - thus too large.
2: It is a Bragg Gray chamber meant to be in equilibrium within the
appropriate phantom. NOT free in air or against metal.
3: Against the film plate is NOT the appropriate position for ENTRANCE
exposure.
Besides - what would be the point of measureing the full integrated R for
the full orbit?? That would really be the DOSE! It WOULD be the entrance
exposure - BUT the insult is not all to the same tissue - thus one would
have to carefully consider the meaning - if any - of the measurement!
The "dwell time" on a given element of tissue is very important - and that
would be evaluated at point of entrance exposure.
In real world - it would be an info only measurement since Panoramic units
seem to have little latitude in exposure parameters - so the exposure really
shoudln't change too much over the life of the unit........ I would expect.
Anyone have experience to the contrary on this last point??