[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: XRD Shutter failure
We had a similar problem with the old type gravity shutter mechanisms and
required continuous GM monitors to be placed near sampler chamber cover
which alarms if shutter failure. Bill Fendt U of DE
On Fri, 16 Dec 1994, Rodney Ice wrote:
>
> I have recently experienced an X-Ray Diffraction Unit Shutter
> failure that resulted in an extremity exposure. The shutter was a Charles
> Supper shutter, approximately ten years old. The shutter, depending upon
> the mounting plate, can be used on a variety of X-Ray Diffraction units.
> The shutter failed to close due to lost tension of the spring around the
> solenoid piston. In addition, the "hairline" wire that closes the audible
> alarm warning signal was displaced from its original position needed to make
> contact.
> A survey of other XRD's on campus located one additional Charles
> Supper shutter of approximately the same age. Evaluation of this second
> shutter indicated a loss of spring tension, however not to the degree of the
> shutter that caused the incident. This second shutter also had the audible
> alarm "hairline" wire out of alignment but it was still making periodic
> contact. The user of the equipment had noted in the user logbook that he was
> having problems with the shutter and that it possibly needed cleaning.
> There is a red warning "flag" or "pin" that extends out of the
> shutter mechanism that also indicates that the shutter is open. In both
> cases, our users had molded thin lead sheets around the shutter so that the
> pin was not visible.
> I suggest that you check any XRD units that you have that are
> operated by aging Charles Supper shutters. Specifically check the red
> "flag" or "pin" for smooth, easy and full tension. There should not be any
> free play in the "flag" or "pin" of the shutter. Any free play indicates
> only partial or no closure of the shutter. Futhermore, check with users
> regarding the audible signal. Is the signal clear and non-varying in
> intensity?
> Although both symptoms (flag and audio) appear to be independent
> problems, both are indicative of aging shutters that should be replaced.
>
> Rodney Ice
> Nuclear Engineering, Georgia Tech
>
>