[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: Low energy pocket dosimeters
>> Does anyone know of a source for low energy pocket ion chambers
>> (self-reading)? Our purchasing department claims they are no longer
>>available.
Isn't XETEX or DCA still selling them - or surely Nuc Assoc??
>> Alternatively, we need a method for getting real time measurements of the
>> total dose from exposure to a pulsed x-ray system with the following
>> characteristics: 50 nanosecond pulse, 25 pulses per second, 150 kVp and 1
>> microseivert per pulse at 1 meter (according to the manufacturer).
CAREFUL!!! I've been there!! 200 mR low energy dosimeters are about a
factor of 2 out of saturation at 10^4 R/hr!!! With that duty cycle you may
be surprised to find how easily you can drive such a dosimeter out of
saturation. Do the quick calculation!!!!
I have actually seen this with pulses about twice that long once a minute or
so. You should check this out, once you get the dosimeters, by comparison
to TLD or film dosimeter - neither of which seem to have a rate problem with
radiation. (yes I know film has a reciprocity problem - but that is only
with light NOT ionizing radiation).
I recomend using the HPI 1030 ion chamber survey meter to measure the dose
rates - it IS expensive - but the only survey instrument I know of that can
maintain saturation at high rates (unfortunately the trade off is that it is
geotropic). Alternative you could calibrate a silicon diode or FAST PM tube
as a detector - but both with me VERY energy sensitive.
>>The dosimeters are to be used by nontechnical personnel (security guards)
>>and are not intended to be record dosimetry. Our primary concerns are
>>expense and ease of use.
LANL developed a CdTe real time dosimeter that MAY be fast enough - but I
don't know if it ever became commercial - AND the room temperature noise of
CdTe is about equivalent to 30 keV photons - thus limiting the low energy
response.
You MAY be between a rock and a hard place here.
>>In a related issue, virtually the only instruments we have found that
>>could detect the radiation emitted by this unit are integrating ion
>>chambers.
I doubt that any of these in any useful range are in saturation either!
Ted de Castro
tdc@ehssun.lbl.gov
University of California Lawrence Berkeley Laboratory
Bldg B75B Rm 101
Berkeley, CA 94720
(510) 486-5256
(510) 486-7304 - FAX