[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: Sterilization of Ir-192 Seeds in Nylon Tubing
Bleach (Sodium Hyperchlorite) is not an acceptable procedure for
sterilization of any items for medical use. Usually ethylene oxide is
the standard procedure for non-metalic items. This is ususally
accomplished with a gas sterilization autoclave. You might possibly
consider steam autoclaving sterilization. I have steam sterilized
nylon tubing using steam autoclaving. I would recommend that you
determine what pressure, temperature is need to render the HIV virus
inactive. There are maybe other chemical available to perform
chemical sterilization, however, continue use of these products will
eventually degrade the nylon tubing. Steam autoclave would be a
possible alternative, however, continue use of this method will
eventually degrade the tubing also. Any major medical facility should
have a ethylene oxide autoclave. This is usually an essential
standard equipment necessary to support operating room suites.
mike coogen sends
______________________________ Reply Separator _________________________________
Subject: Sterilization of Ir-192 Seeds in Nylon Tubing
Author: radsafe@romulus.ehs.uiuc.edu at hq2ccgw
Date: 8/29/95 10:07 AM
Has anyone ever had to sterilize nylon tubing containing Ir-192 seeds for an
HIV positive patient? Our Radiation Oncology dept. wants to use bleach, but
there is concern that it will degrade the nylon. The company who
manufactured the seeds says that they should only use gas sterilization, but
for several reasons this is not currently an option.
Any information would be greatly appreciated.
-------------------------------------------------------------------------
Melanie Byers, Health Physicist Vanderbilt Univ. Institutional Safety
byersmm@ctrvax.vanderbilt.edu Radiation Safety Section
(615)322-2057 (phone) Nashville, TN
-------------------------------------------------------------------------
This is my opinion only and does not necessarily reflect my employer's.