[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Q: "Tandetron" accelerator and regs
From: shand
Hi Melissa:
We have a "Pelletron" 1.7 Mev accelerator here at
UMCP. We are also an agreement state. Specific authorized
use at this time (in house under our license) is for hydrogen
and helium ions as well. Our state regulations call for
an interlock: "Each entrance into a target room ... shall be
provided with a safety interlock that shuts down the machine
under conditions of barrier penetration." There are two
entrances to the room, each is interlocked. The console is
located about 5 feet from the target area, in the same room.
The initial run found a 1000 mr/hr area between the plasma
source and the bending magnet. This area we wrapped in lead
and reduced to less than 30 mR/hr on contact. At one meter
from the tank the exposure rate was 0.05 mr/hr. The highest
reading on the tank was 0.2 mr/hr.
A Dr. Correl at the US Naval Academy in Annapolis, Maryland
was very helpful with our understanding of the machine. We
visited a 2 Mev Pelletron there and observed the warning
light operation, door interlocks testing and measurments
while in operation. Their console was also in the room
and located about 20 feet from the target area.
If you would like, I can fax a schematic of our machine, the
exposure rates at various points, etc...
Hope this was helpful
steve hand
UMCP