[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: Thallium 201
Darryl Kaurin wrote:
>
>Dear Radsafers,
>
>Regarding the stress test, I had one and they did use 4 mCi of thallium-201.
>The health physics personal where I worked measured me daily, and the half
>life was 3 days, equal to the physical decay. I couldn't work in radiation
>areas for 3 weeks. I was told by the hp staff latter that you might be able
>to request they use Tc-99m insted, which has a much shorter half life, and
>less dose of course too. I haven't calculated my dose, but it is certainly
>more than the dose I get at work.
>
Well, perhaps less dose. These are the doses, in mGy/MBq of Tl-201 and
Tc-99m sestamibi, in resting patients or patients undergoing stress testing,
and the effective dose equivalents and effective doses in mSv/MBq:
Tl-201 MIBI - rest MIBI - stress
Adrenals 6.3E-02 4.3E-03 3.9E-03
Brain 5.9E-02 1.8E-03 1.9E-03
Breasts 3.6E-02 1.7E-03 1.6E-03
GB Wall 8.3E-02 1.8E-02 2.5E-02
LLI Wall 3.4E-01 3.7E-02 2.9E-02
Small Intestine 4.5E-01 2.7E-02 2.2E-02
Stomach 1.9E-01 5.2E-03 4.7E-03
ULI Wall 3.3E-01 5.0E-02 4.0E-02
Heart Wall 2.8E-01 4.4E-03 4.8E-03
Kidneys 4.6E-01 1.8E-02 1.5E-02
Liver 9.9E-02 5.1E-03 3.7E-03
Lungs 4.7E-02 2.4E-03 2.2E-03
Muscle 4.6E-02 3.7E-03 3.3E-03
Ovaries 1.0E-01 1.4E-02 1.2E-02
Pancreas 7.4E-02 5.0E-03 4.6E-03
Red Marrow 5.5E-02 4.5E-03 4.0E-03
Bone Surfaces 8.8E-02 5.8E-03 5.4E-03
Skin 3.3E-02 1.9E-03 1.8E-03
Spleen 1.8E-01 5.2E-03 4.0E-03
Testes 2.0E-01 3.5E-03 3.1E-03
Thymus 4.6E-02 2.3E-03 2.3E-03
Thyroid 6.2E-01 2.2E-03 2.2E-03
Urinary Bladder Wall 5.2E-02 3.7E-02 2.7E-02
Uterus 8.5E-02 1.2E-02 1.0E-02
Effective Dose Equivalent 1.6E-01 1.5E-02 1.3E-02
Effective Dose 1.4E-01 1.3E-02 1.1E-02
Bear in mind, however, that a typical Tl-201 test uses only around 50 MBq,
while a Tc-99m MIBI test may use 500-1000 MBq. So the radiation doses are
fairly comparable. In addition, some physicians will use multiple
injections of Tc to study the initial uptake, and later, the redistribution,
while Tl-201 has very good inherent redistribution. OTOH, the Tc-99m is
better than Tl for doing first pass studies, which may be done in
conjunction and save the additional administration of a separate
pharmaceutical. But my point is just that the dose picture is not that
markedly different. I also thought that an accurate picture of the
dosimetry would be interesting. You can always get access to such numbers
at our web site.
I also agree with Liz Brackett's comment about the presence of contaminants
and their effect on external exposure. The numbers above for Tl-201 include
these contaminants, BTW.
Mike Stabin
Radiation Internal Dose
Information Center
Oak Ridge, TN
http://www.orau.gov/ehsd/ridic.htm