Ken,
I use a Flash X-ray machine that is in the 3 MeV range with a pulse
duration of 20 nsec. Yes sheilding is very much required on these
machines. The charging potential on our machine is 30KV and the total
dose would be approximately 4700 R or 2.35 E11 R/sec (with a 20 nsec
pulse duration).
Please feel free to contact me at kemp@plk.af.mil
Bill Kemp
RSO for Phillips Lab/VT
______________________________ Reply Separator _________________________________
Subject: Flash X-rays
Author: radsafe@romulus.ehs.uiuc.edu at smtpgw1
Date: 6/11/96 12:17 PM
I received a request for information about radiation protection requirements
for use with 'flash' x-rays. The gentleman who called said this would use
x-ray plses in the MeV range with a duration of less than a microsec per
pulse but with enough intensity to produce an exposure of normal density. My
reaction was that the requirements would depend on the total radiation
exposure (or workload) similar to a normal x-ray device. He said he
remembered vaguely from some previous work that such short pulses somehow
reduced the need for shielding, perhaps by producing less damage in the cell.
Does anyone have any experience with flash x-rays or know of any special
protection requirements?
Thank you for your help.
Ken
Ken Douglass, Ph.D.
Interim Radiation Safety Officer
West Virginia University Hospitals, Inc.
West Virginia University
P.O.Box 8150
Morgantown WV 26506
kendoug@wvnvm.wvnet.edu