Julia Sober wrote:
>
> Dear Radsafers,
>
> I have recently obtained a Ludlum Model 3/44-9 and am in the process of
> performing efficiency calculations for various isotopes in use at my
> facility, namely P-32, S-35, I-125, and C-14. In determining efficiency, I
> plan to take a measured amount of each isotope, place it on a disc
> smear, determine the count rate, and perform the necessary
> calculations.
>
> Yesterday the above method gave me a 30.6% efficiency for P-32.
>
> I would like to know if this is within the expected range for this
> instrument and if this is an efficient (so to speak) method for determining
> efficiency.
>
> I would also appreciate a list of expected efficiency ranges for the
> Model 3/44-9 for the isotopes I have listed if anyone is using similar
> instrumentation and has these numbers handy.
>
> In Anticipation,
> Julia Sober
> Acting RSO
> Northern Illinois University
> DeKalb, IL 60115
>
> jsober@niu.edu
> Phone (815) 753-1093
> Fax (815) 753-0461
Dear Julia and Radsafers,
Ludlum Measurements, Inc. lists the counting efficiency as 2-pi rather
than 4-pi. I believe most people mean 4-pi counting efficiency when
they speak of counting efficiency. Thus, if Ludlum gave a figure of
60%, you would expect to measure about 30%.
There is even more confusion in the simple term "counting efficiency"
when the abundance of emission is different from unity.
I recommend that we all use the term "counts per disintegration" instead
of "counting efficiency". For a specified rationuclide, c/d is fully
defined and without confusion. Any comments?
cheers, Wes
--