Luminescence & LSC
Joel Baumbaugh recently posted a message about LSC chemiluminescence in
which he stated that such was a single photon event and not a problem with
newer LSC counters. This has not been my experience.
I have never heard of this "single photon" bit, but I do know that chemo is
still a major concern, at least for environmental samples (e.g., lab wipes).
Perhaps a vigorous reaction emits several photons from several foci that
seem to the LSC as a single real event. In any case, we still see it even
with the latest cocktails and counters. Several popular bench cleaners are
a real pain. There are three things you can do to counter chemo:
1) Chemo appears in the lower half of the H-3 window. You can count in
the full window and then crank the upper level discriminator down to 50% and
count again. A drop in the count rate of about 50% means H-3. No change
means chemo
2) If your counter can print a spectrum, do that and see if all the
counts are in the lower half of the H-3 range (chemo).
3) Chemo has a much shorter half-life than H-3. If you have the time,
count your sample in the morning and again in the evening. A significant
drop in the count rate means that it is chemo. It has been our experience
that all chemo disappears after 3 days. For our lab wipes of H-3 labs, we
simply fill the vials and then let them sit for 3 days and count then.
Whatever we see after that we assume to be H-3.
**********************************************************************