For a bone scan, in which the standard dose is 20 mCi (740 MBq), the
radiation absorbed dose in rad has been estimated as follows to a
standard adult person (ref:Practical Nuclear Medicine; Palmer, Scott
and Strauss, W.B. Saunders & Co, 1992):
total body 0.13
red marrow 0.56
bladder wall 2.6 ( 2 hour void)
6.2 ( 4.8 hour void)
ovaries 0.24 ( 2 hour void)
0.34 ( 4.8 hour void)
testes 0.16 ( 2 hour void)
0.22 ( 4.8 hour void)
Mike Quastel
Inst. of Nuclear Medicine
Soroka Medical Center
POB 151, Beer Sheva, Israel 84101
----------------------------------------------------------------
On Wed, 13 May 1998, Ruth Weiner wrote:
>
> I recently had a bone scan and was injected with Tc-99. Using my
> handy Pink Book (now a Blue Book) I calculated the EDE to be about 60
> mrem.
>
> Clearly only my own opinion.
>
> Ruth Weiner
> rfweine@sandia.gov
>
>
> ______________________________ Reply Separator _________________________________
> Subject: Re: EDE from FDG
> Author: alstonc@odrge.odr.georgetown.edu at hubsmtp
> Date: 5/13/98 6:41 AM
>
>
> Scott
>
> It's not, however, considerably greater than the EDE from other nuc med
> diagnostics, but, in fact, readily comparable. Somehow, it seems intuitively
> "right" that it takes a bigger dose to get physiological data, than to get
> anatomical data, doesn't it?
>
> chris alston
>
> At 10:41 05/12/98 -1000, you wrote:
> >I was recently asked to compare the dose received from a PET exam to a
> >chest x-ray. The best reference I found was in the Journal fo Nuclear
> >Medicine 1991; 32:699-706. In Table 6, the EDE for an intravenous
> >injection of FDG is stated as 2.4E-2 mGy/MBq.
> >
> >That would be 0.09 rad/mCi in nostalgic units. Considering a typical PET
> >study may require 10 mCi of FDG, that means the EDE would be 0.9 rad,
> >or 900 mrem.
> >
> >I was very surprised by the magnitude of this value. Table 3.27 of NCRP
> >100 states the EDE of a chest exam to be 0.08 mSv, or 8 mrem. The EDE
> >from a CT exam is stated as 1.11 mSv, or 111 mrem. The EDE from a
> >barium enema exam is 4.06 mSv, 0r 406 mrem.
> >
> >As you can see, the EDE from a FDG PET exam is considerably greater
> >than other diagnostic exams. This was news to me, so I thought it might
> >be interesting to share with you.
> >
> > *****
> >Frequently Asked Questions (FAQ) for Medphys at: