Good Point! - I'd like to add a little to this.
You should correct for "backscatter" if you use a metal backed, beta
calibration source, especially if you will be using the instrument for a
sample which may have a different backscatter factor, such as a smear or air
sample. Figure 2-17, of Glenn Knoll's book, "Radiation Detection and
Measurement", (New York, Wiley, 1989) includes backscatter fractions for
various electron energies and materials. (If you haven't guessed this by
now,
I'm a big fan of this book. Every HP should have a copy!) Note that, per
Glenn, "Backscattering is most pronounced for electrons with low incident
energy and absorbers with high atomic number." There is also a useful graph
in the 1970 edition of the "Radiological Health Handbook", page 127.
I've found that the metal used for the source backing will have a
significant
impact on the apparent efficiency, even for metals with similar backscatter
factors, eg. stainless steel and nickel. It's probably due to the way the
radioactive material adheres to the metal, leading to different degrees of
self shielding. If you participate in a cross check program where the cross
check standards are a different metal than your calibration sources, this
may
result in an apparent systematic error.
One final point and I'll leave you alone. Use a great deal of caution if you
use a windowless counter for assaying nonconducting sources, such as smears
or
air samples. Static charge buildup may lead to nonreproducible results.
The opinions expressed are strictly mine.
It's not about dose, it's about trust.
Bill Lipton
liptonw@dteenergy.com
You wrote:
>As a caveat - One can get surprisingly good efficiency (close to 50%) with
high energy >betas (e.g. P-32 and/or Sr-90) if you are on a hard flat
surface
because the betas >actually bounce back from the surface into the detector
(enhancing the response). >This would not be the case with a check source
which would contain the P-32 (you only >get about 35% eff.) but with a
contaminated surface like a lab-bench top which >could/would/will interact
with the surface you get more "bang for your buck"...
************************************************************************