[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
GM Response
- To: RADSAFE <RADSAFE@romulus.ehs.uiuc.edu>
- Subject: GM Response
- From: "John R. Laferriere 671-8316" <John.R.Laferriere@dupontpharma.com>
- Date: Tue, 08 Sep 1998 10:29:28 -0400 (EDT)
- A1-type: MAIL
- Alternate-recipient: prohibited
- Hop-count: 1
- Importance: normal
- MR-Received: by mta LLVX02; Relayed; Tue, 08 Sep 1998 13:04:15 -0400
- Posting-date: Tue, 08 Sep 1998 13:04:00 -0400 (EDT)
- Priority: normal
- UA-content-id: D2589ZXOBIOYWT
- X400-MTS-identifier: [;51403180908991/5289440@LLDMPC]
Radsafers:
Although we always use LSC for "official" wipe surveys
in labs handling Ni-63, we also use a Ludlum model 120
gas flow proportional counter with 0.4 mg/sq.cm. window
for initial in-lab screening of daily operational wipes by
the lab workers, and have also discovered that a pancake
GM can be helpful for gross screening for hot spots in
the lab. (The lab handles GBq's of Ni-63 on a regular
basis.)
The gas flow was calibrated with a traceable Ni-63 solution
deposited on a plastic substrate to minimize self-absorption,
and is normally in the 0.01 c/d efficiency range. For wipes,
the efficiency is about half that determined by the standard,
the drop being due to self-absorption. We use a scaler with
the model 120, and counting times of a minute or less, for
gross screening.
We recently cross calibrated a pancake GM with the LSC
and a hot wipe, and got about 0.0018 c/d (0.18%) on
the pancake. Given the lab background, I estimated about
a 50,000 dpm MDA for the pancake/ratemeter combination,
using a near-contact geometry. Not a great efficiency,
but better than the zero % that we were expecting, and
good enough to know if there's a real problem developing.
And being able to monitor work surfaces directly is a real
help for lab workers, compared with semi-blind wipes.
Years ago at another company, we used a thin NaI "low energy
probe" to roughly assay mCi quantities of H-3 in a skin puncture
wound, via detection of bremsstrahlung. I don't recall the
efficiency, but the fact that it could be done at all was
interesting.
John R. Laferriere, C.H.P.
Medical Imaging Division,
DuPont Pharmaceuticals Company
john.r.laferriere@dupontpharma.com
(978) 671-8316
************************************************************************
The RADSAFE Frequently Asked Questions list, archives and subscription
information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html