[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
RE: Survey Instrument calibration frequency
Thank you everyone who has responded so far, particularly to Chris Wend and
Bruce Busby, who reminded me to pull NCRP 112 off my shelf. I looked at it 6
years ago when I was renewing my license LAST time. For license purposes, we
use ANSI N323 because it's shorter. I like Appendix Q on NUREG 1556 vol. 11
because it's shorter still.
Anyway, NCRP 112, chapter 7 says: "The underlying reasons for performing
surface activity measurements often do not necessitate particularly accurate
measurements.... For purposes of common radiation control, routine
measurements of surface contamination are made to fulfill regulatory
requirements and to provide semiquantitative information on which to base
further action (e.g., decontamination). Under such circumstances, a
sophisticated and time-consuming calibration of a monitoring instrument is
not justified." What's it mean by that? Any comments? I'm investigating
further.
I'm trying to do what I believe is the right thing, without burying myself
with my own good intentions. What do I mean by that? A few years ago we were
cited by NRC because our geiger counters were not calibrated within six
months. This calibration frequency was our license condition, not based on
ANSI, NCRP, or any other recommendation. The old RSO just did it that way. I
soon changed that. Now I follow ANSI N323-1978, but I want to upgrade to the
newer model (or an acceptable alternative).
Jim Herrold
Radiation Safety Officer
University of Wyoming
herrold@uwyo.edu
-----Original Message-----
From: Christopher_Wend@bedison.com [mailto:Christopher_Wend@bedison.com]
Sent: Thursday, February 04, 1999 12:09 PM
To: Multiple recipients of list
Subject: Re: Survey Instrument calibration frequency
Brief summary:
ANSI N323A-1997, section 4.9, says: "Calibration shall be required at least
annually..." There are also the usual words about calibration after
maintenance,
adjustment, etc. that can affect instrument performance. Also words about
increasing cal. frequency due to extreme op, conditions, hard usage, etc.
Essentially, there isn't any change (on this issue) from the 1978 version.
NCRP 112, section 2.8 begins with: "Calibration frequencies suggested by
various
groups for portable radiological survey instruments vary from once every few
weeks to once every year, with the most commonly suggested periods between
calibrations ranging between quarterly and annually."
ANSI N323A-1997, NCRP 112 and ISO 7503-1 all require daily source response
checks (NCRP calls it a "performance check"). The source response checks
should
be based on a reference reading, normally established during calibration.
Acceptance criteria should be +/-20% (ANSI N323A-1997, section 4.8).
I'd be wary about taking your cal. freuency past 1 year. Recommend reviewing
the
discussion in NCRP 112, section 2.8. Probably not what you wanted to hear -
but
I hope it helps a little.
My opinion only,
Chris Wend
Sr. Rad Engineer
(508) 830-8158
christopher_wend@bedison.com
magnum8@banet.net
______________________________ Reply Separator
_________________________________
Subject: Survey Instrument calibration frequency
Author: "Jim F. Herrold" <Herrold@uwyo.edu> at Internet
Date: 2/4/99 12:27 PM
Howdy.
Reading through Draft NUREG 1556 vol.11 I came across this statement under
section 8.10.2, RADIATION MONITORING INSTRUMENTS:
"Some instruments may only need to be checked periodically for operability
and response to radiation rather than receive full calibration. For
example, Geiger-Mueller (G-M) type survey instruments used to identify
contamination in laboratories may only need to be checked for ability to
detect low level contamination."
If this is true, it would have a HUGE impact on our calibration program. Is
anybody else doing this? Does anybody know where the NRC got this guidance?
Was it from ANSI N323A-1997? I don't have it yet, but am planning to obtain
one as soon as possible. We have been following the old ANSI N323-1978, and
don't remember any statements like the one above.
We are in the process of renewing our license. Because NUREG 1556-11 is
still in draft form, the NRC says I can't use it yet. However, if I can find
another suitable reference for this statement, I could use it.
Thanks.
Jim Herrold, Radiation Safety Officer
herrold@uwyo.edu
University of Wyoming
Environmental Health & Safety
303 Merica Hall
Laramie, WY 82071-3413
(307) 766-3277
************************************************************************
The RADSAFE Frequently Asked Questions list, archives and subscription
information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html
************************************************************************
The RADSAFE Frequently Asked Questions list, archives and subscription
information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html
************************************************************************
The RADSAFE Frequently Asked Questions list, archives and subscription
information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html