  Dear Radsafers,
              I wonder if some of you could suggest an explanation for the
following problem I have. It concerns the considerable differences I am
finding between the  expected and measured dose in the main beam of an
X-ray source.
              The tube is running at 60 kV with a tube current of 3 mA.
The filtration is 1 mm Be. According to my calculations, at 17cm from
the window, the dose should be 170 cGy (170 rads) for a 10 sec exposure.
A TLD in the main beam consistently records a factor of about 10x less
i.e. 140 mGy (14 rads). A series of measurements under different
conditions reproduces this factor of a 10x inconsistency. Would this be
expected? Should I not be using a TLD for this type of measurement and if
not, what should I be using?
     Any suggestions will be most welcome.

                 David Hornsey


        

 **************************************
 * David J. Hornsey, Radiation Safety.*
 * South Building,                    *
 * University of Bath                 *             
 * Claverton Down,                    * 
 * Bath BA2 7AY. UK.                  *
 *                                    *
 * tel:01225 826540                   *
 * fax:01225 826779                   *  
 * e-mail: d.j.hornsey@bath.ac.uk     *
 **************************************

************************************************************************
The RADSAFE Frequently Asked Questions list, archives and subscription
information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html