Joseph Guido wrote:
I believe that the issue here is the error associated with the net count
rate. It has been a while since I looked at an E-140, but I assume it is
being used as a ratemeter in this application.According to Eberline, the E-140's response time is variable from 2 to 10
seconds. So, not knowing how it is set up, look at both extremes -
keeping in mind that the response time is equal to 2.3 time constants
(according to Eberline).Note that these calculations assume Poisson statistics:
error (1 sigma) = sqrt (count rate / (2 * time constant)) :Cember 3ed p403
time constant = response time / 2.3 : Eberline tech support
In the high background application-----------------
for a 2 second response time -
a 1500 cpm gross reading has an associated 1 sigma error rate of 227
cpm.for a 10 second response time -
a 1500 cpm gross reading has an associated 1 sigma error rate of 102
cpm.In the low background application---------------------
for a 2 second response time -
a 550 cpm gross reading has an associated 1 sigma error rate of 138
cpm.for a 10 second response time -
a 550 cpm gross reading has an associated 1 sigma error rate of 62
cpm.----
Going one step further, you would express the net count rate as:
net = gross -bkg +/- sqrt (error_gross^2 + error_bkg^2)
In the high background application
gross count rate = 1500 cpm
bkg count rate = 500 cpm500 +/- 294 cpm (2 second time constant)
500 +/- 131 cpm (10 second time constant)In the low background application
gross count rate = 550 cpm
bkg count rate = 50 cpm500 +/- 164 cpm (2 second time constant)
500 +/- 64 cpm (10 second time constant)---
joe guido