I have used the Ludlum 43-37 extensively in the past for floor surveys in my
previous job. Here's some info:
-Purge time approx. 60 min (approx. 60 cc/min.)
-HV approx. 1800v
(This was determined by generating a voltage vs. response with a Tc-99 check
source.)
-We used a Ludlum Model 500 pulse generator to calibrate all of our meters.
This is good for checking linearity and the accuracy of the scaler feature.
-We generally had 3-5% efficiency to C-14 (probe at 1 cm from source)
These are just a few brief notes for more detail rely to tmikell@excite.com
Toli Mikell
On Thu, 25 May 2000 11:19:54 -0500 (CDT), radsafe@romulus.ehs.uiuc.edu
wrote:
>
> I need to create a operation/calibration procedure for a Ludlum 43-37
floor
> monitor. I'll probably use the unit in conjunction with Ludlum 2221
meters (one
> set up for alpha and another set up for beta/gamma). I would appreciate
any
> advise or preferably copies of procedures from other users of this or
> similar-type floor monitor.
>
> Thanks in advance,
>
> Rodney Bauman, CHP, RRPT
> Weldon Spring Site Remedial Action Project
> rodney_bauman@wssrap-host.wssrap.com
> (636) 441-8086 ext. 4219
> fax: (636) 926-7072
>
> ************************************************************************
> The RADSAFE Frequently Asked Questions list, archives and subscription
> information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html
_______________________________________________________
Get 100% FREE Internet Access powered by Excite
Visit http://freelane.excite.com/freeisp
************************************************************************
The RADSAFE Frequently Asked Questions list, archives and subscription
information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html