I think it is informative to get more specific details on this incident. I
was surprised at the 3-15 rem dose level and could not understand how that
could happen but then when I found out the adminstered activity was 285
mCi.... The notice of violation is at:
http://www.nrc.gov/reading-rm/doc-collections/enforcement/actions/materials/
ea02248.html
The interesting paragraphs are below. And yes, I think the RSO's feet should
be held to the fire. I don't think the NRC's expectations are unreasonable.
Hobie Shackford
Roger Williams Medical Center
Providence, RI
Specifically, for the period from July 1 through July 7, 2002, a member of
the public received a total effective dose equivalent between 3 and 15 rem.
The individual received the exposure while visiting a relative who had been
administered a radiopharmaceutical therapy dosage of 285 millicuries of
sodium iodide iodine-131. The individual received the exposure as a result
of not following prescribed radiation safety instructions, including staying
behind the shielding. Hospital staff frequently noted the individual sitting
at the patient's bedside where a shield was not located. When this was
observed, the hospital staff reminded the individual to position herself on
the other side of the bed behind shielding. While the member of the public
disregarded radiation safety instructions provided by hospital staff, the
radiation safety officer failed to implement corrective actions as necessary
to minimize the individual's exposure, once the extent of the potential
exposure was identified.
During the predecisional enforcement conference, you stated that you did not
take more proactive steps to limit the dose to visitors or to contact the
patient's relatives to more accurately determine their dose because of your
concern for the patient's rights and for compassion toward the patient and
family members. The NRC staff understands that the case involved unusual
circumstances and that you did not want to appear uncompassionate toward the
patient or her family. However, you could easily have implemented a number
of actions in response to the daughter not following instructions, such as:
(1) explaining to the daughter that staying an arm's length from the patient
would significantly reduce the exposure (radiation levels at one meter were
approximately one-tenth those at the bedside); (2) using additional
shielding, including shielding the catheter bag; (3) minimizing the
daughter's time at the bedside; and (4) providing a digital dosimeter for
the daughter to self-monitor her exposure, which you had available.
Therefore, the NRC has determined that your staff's performance was
deficient such that enforcement action is warranted.
Although the NRC's medical consultant indicated that the radiation exposure
received by this individual is not significant from a health and safety
standpoint, the NRC considers any exposure in excess of regulatory limits a
significant matter. The NRC expects licensees to conduct their operations in
a manner that precludes such exposures from exceeding the limits established
in 10 CFR Part 20. In this case, the exposure to the individual was well in
excess of the regulatory limit. Therefore, these violations are categorized
collectively in accordance with the "General Statement of Policy and
Procedure for NRC Enforcement Actions" (Enforcement Policy), NUREG-1600 as a
Severity Level I problem.
************************************************************************
You are currently subscribed to the Radsafe mailing list. To unsubscribe,
send an e-mail to Majordomo@list.vanderbilt.edu Put the text "unsubscribe
radsafe" (no quote marks) in the body of the e-mail, with no subject line.
You can view the Radsafe archives at http://www.vanderbilt.edu/radsafe/