Re: Rad Safe Blood Irradiator
Laura-
        The flux reduction could be due in part to irregularities/pits in 
the anode-  a smooth surface will produce the most, 
( i.e. when new),   a roughened surface caused by normal wear will reduce 
the output......  But I'd bet that the real culprit is the gradual 
evaporation- and subsequent deposition- of the Tungsten filament upon the 
tube's window.   ( you mentioned glass, it might be beryllium?)   It 
doesn't take much  W  to reduce the flux, particularly the softer part of 
the tube's spectrum.  The harder you run the tube ( higher mA /filament 
power) the faster the W will boil off the filament.   If consistent flux 
rates are important,  you may have to consider the tube to be a 
'consumable' ......    Depending on what type of tube it is, ( fixed 
target/rotating anode) there may be some chance that the tube can be 
rebuilt. 
        Good Luck!
                Bob Westerdale
Hello,
I would like to request a RADSAFE brain trust on an 
issue we have with our Rad Safe Blood Irradiator.
We purchased a New Rad Safe Blood Irradiator in 2001 - which was at the 
time a new technology that uses
an X-ray Tube versus the standard Cs-137 for irradiating Blood products. 
Below I have listed our service
schedule, but the summary is every 6-12 months we are forced to increase 
the time the of irradiation due
to our TLD dosimetry measurement falling below the required 2500cGy+_10% 
we require for our standard.  The time has
had to be increased to such a length that it will soon perhaps interfere 
with our duties and blood irradiation flow rate.
The manufacturer says they are not sure what is causing this and are 
currently performing research to 
determine why time always needs increasing.  My initial thought was 
flaking of the anode, causing a filter effect
on the glass which continuous increases and filters out radiation, 
therefore taking longer to get to the 2500cGy 
value.  The manufacturer says they do not think so based on research, but 
are unsure - they do not manufacturer the
tubes and are in consult with the tube manufacturer. 
Other than the flaking of the anode theory, what might cause this?
Laura Smith
Date                                             Issue           TLD Dose  
 Time
8/8/2001                                 New unit                2534   
8/10/2001                                independent evaluator   2734   
6.4min
9.27/01                                          Dose check   2350   ""
10/19/01                                 Qtrly Dose Check        2660   ""
1/24/02                                          ""   2610       ""
5/24/02                                          ""   2460       ""
7/12/02                                          ""   2440       "" - time 
increased
10/10/02                                 time increased check again  2460  
                 6.8
12/26/02                                 Annual service contract   2360    
6.8
1/15/03                                          Qtrly Dose Check   2190   
                 6.8 - time increased
2/4/03                                           time increased check 
again            2530                            7.5
4/7/03                                           Qtrly Dose Check   2360   
                 7.5 - time increased 
7/9/03                                           time increase check again 
         2460                            8.1
CONFIDENTIALITY NOTICE:  This email message and any accompanying data are 
confidential, and intended only for the named recipient(s).  If you are 
not the intended recipient(s), you are hereby notified that the 
dissemination, distribution, and or copying of this message is strictly 
prohibited.  If you receive this message in error, or are not the named 
recipient(s), please notify the sender at the email address above, delete 
this email from your computer, and destroy any copies in any form 
immediately.
************************************************************************
You are currently subscribed to the Radsafe mailing list. To unsubscribe,
send an e-mail to Majordomo@list.vanderbilt.edu  Put the text "unsubscribe
radsafe" (no quote marks) in the body of the e-mail, with no subject line.
You can view the Radsafe archives at http://www.vanderbilt.edu/radsafe/