[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]

Re: Rad Safe Blood Irradiator



Laura-

        The flux reduction could be due in part to irregularities/pits in 

the anode-  a smooth surface will produce the most, 

( i.e. when new),   a roughened surface caused by normal wear will reduce 

the output......  But I'd bet that the real culprit is the gradual 

evaporation- and subsequent deposition- of the Tungsten filament upon the 

tube's window.   ( you mentioned glass, it might be beryllium?)   It 

doesn't take much  W  to reduce the flux, particularly the softer part of 

the tube's spectrum.  The harder you run the tube ( higher mA /filament 

power) the faster the W will boil off the filament.   If consistent flux 

rates are important,  you may have to consider the tube to be a 

'consumable' ......    Depending on what type of tube it is, ( fixed 

target/rotating anode) there may be some chance that the tube can be 

rebuilt. 

        Good Luck!

                Bob Westerdale











Hello,

I would like to request a RADSAFE brain trust on an 

issue we have with our Rad Safe Blood Irradiator.



We purchased a New Rad Safe Blood Irradiator in 2001 - which was at the 

time a new technology that uses

an X-ray Tube versus the standard Cs-137 for irradiating Blood products. 

Below I have listed our service

schedule, but the summary is every 6-12 months we are forced to increase 

the time the of irradiation due

to our TLD dosimetry measurement falling below the required 2500cGy+_10% 

we require for our standard.  The time has

had to be increased to such a length that it will soon perhaps interfere 

with our duties and blood irradiation flow rate.



The manufacturer says they are not sure what is causing this and are 

currently performing research to 

determine why time always needs increasing.  My initial thought was 

flaking of the anode, causing a filter effect

on the glass which continuous increases and filters out radiation, 

therefore taking longer to get to the 2500cGy 

value.  The manufacturer says they do not think so based on research, but 

are unsure - they do not manufacturer the

tubes and are in consult with the tube manufacturer. 



Other than the flaking of the anode theory, what might cause this?



Laura Smith







Date                                             Issue           TLD Dose  

 Time

8/8/2001                                 New unit                2534   

8/10/2001                                independent evaluator   2734   

6.4min

9.27/01                                          Dose check   2350   ""

10/19/01                                 Qtrly Dose Check        2660   ""

1/24/02                                          ""   2610       ""

5/24/02                                          ""   2460       ""

7/12/02                                          ""   2440       "" - time 

increased

10/10/02                                 time increased check again  2460  

                 6.8

12/26/02                                 Annual service contract   2360    

6.8

1/15/03                                          Qtrly Dose Check   2190   

                 6.8 - time increased

2/4/03                                           time increased check 

again            2530                            7.5

4/7/03                                           Qtrly Dose Check   2360   

                 7.5 - time increased 

7/9/03                                           time increase check again 

         2460                            8.1









CONFIDENTIALITY NOTICE:  This email message and any accompanying data are 

confidential, and intended only for the named recipient(s).  If you are 

not the intended recipient(s), you are hereby notified that the 

dissemination, distribution, and or copying of this message is strictly 

prohibited.  If you receive this message in error, or are not the named 

recipient(s), please notify the sender at the email address above, delete 

this email from your computer, and destroy any copies in any form 

immediately.

************************************************************************

You are currently subscribed to the Radsafe mailing list. To unsubscribe,

send an e-mail to Majordomo@list.vanderbilt.edu  Put the text "unsubscribe

radsafe" (no quote marks) in the body of the e-mail, with no subject line.

You can view the Radsafe archives at http://www.vanderbilt.edu/radsafe/