[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
RE: Puzzler: How Does Humidity Affect Counting Efficiency?
Had the service rep in on Friday to check out the Tri-Carb. I showed him
the data and went over the Radsafe feedback with him on the tritium check
source results proportional to humidity. First, he said that he had never
heard of anyone having this problem before. Of course, he also has never
heard of anyone calibrating at 10% humidity and operating at 60% humidity!
He called the Technical Expert at the main office and relayed his thoughts
on what might be happening. They think that air humidity is condensing on
the glass vials while they're in the refrigerated Tri-Carb cabinet. The
water on the outside of the glass is luminescencing while it's counting.
According to the expert, it's easier for the Tri-Carb to discriminate
non-tritium counts at lower count rates than at higher count rates. I'm
not sure of the technical reason for this.
One guess from me: the Tri-Carb has two PM tubes and the sample is counted
between them. It may be that it by having the PM tubes simultaneously
counting the sample, counts coming from non-uniform condensed humidity on
the outside of the vial do not match the timing criteria that it expects to
see for tritium counts. Of course, I would think that even if the
condensed humidity on the outside of the vial luminescents, the light
spectrum would be quite different than from the tritium light spectrum
coming from inside the vial and easy to discriminate at any count rate.
The problem in both cases may be that it can't completely discriminate
against the false counts due to spill-over into the tritium portion of the
spectrum. But I still can't figure out why it only happens at higher count
rates.
Anyone have any thoughts on this? Does it seem plausible to you? I do
have to confirm that humidity is making it's way into the cabinet. In very
humid conditions, when we open up the top to put in a sample, it's actually
raining inside (water dripping off the inside of the cabinet).
John M. Sukosky, CHP
Dominion
Surry Power Station
(757)-365-2594 (Tieline: 8-798-2594)
************************************************************************
You are currently subscribed to the Radsafe mailing list. To
unsubscribe, send an e-mail to Majordomo@list.vanderbilt.edu Put the
text "unsubscribe radsafe" (no quote marks) in the body of the e-mail,
with no subject line. You can view the Radsafe archives at
http://www.vanderbilt.edu/radsafe/