[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Bi207
Hey thanks for the responses, there is some difference in answers (i.e.
41mCi - 60mCi). Here is what I did and my results:
I used the 1970 edition of the RHH to get the equation and gamma energies
R/hr @ 1ft = 6 C E n
where C = # Ci's
E = gamma energy in MeV
n = gamma quanta/dis
C = 0.6R/hr / 6En
where 6En = 6 * ((0.57 * 0.98) + (1.1 * 0.77) + (1.8 * .09))
= 6 * (0.5586 + 0.847 + 0.162)
= 6 * 1.5676
= 9.4056
goto original equation
C = 0.6 / 9.4056
= 0.06379
= 63.8 mCi
The instrument that I used was a Victoreen model 470A.
Thanks
Denny Rinkacs
Rad Safety
CMU
PS sorry about the 35mR/hr or 35R/hr, it is in fact reading 35 R/Hr on
contact with the above mentioned instrument.