Re: TLDs - False Positive Betas



We issue about 130000 dosimeters per year and see the false positive on 
occasion.  We are using a 4 element tld with a 0.0036" thick element for Beta 
particle detection.  The other elements are 0.015".  The thin chip is no longer 
sold with that thickness.

A study of this problem using 10,000 cards was performed here at ORNL by Ed 
Sonder and Abu Ahmed.  It is called "Analysis of Anomalous Data Produced by 
Harshaw Model 8801 Thermoluminescent Dosimeter Cards" ORNL/TM-12311, May 1993.  
They found that 218 cards had a problem with the thin chip and only 3 cards had 
a similar problem with a thick chip.

To quote the paper: "These false readings are usually accompanied by irregular 
glow curve shapes.  Once a chip gives bad results, its probabilityu of giving 
future excess readings is elevated.  Moreover, the history of the suspect 
dosimeters often indicates a previous tendency towards elevvated readings for 
the same chip."  You can either call the ORNL library to get a copy, or me.

ORNL Library:  615 574-6744

Good luck,

Rob Gunter