Re: ED underresponse in pulsed x-ray fields
>>However, I have heard from someone at LBL (I think, it's been a while) who
>>used the same dosimeter in a Mega-rad/hr (instantaneous) field (pulsed as
>>well) and they responded admirably.
Well maybe not mega-rad/hr - then again......
We exposed to the highest fields we could find around the ALS - a few 10's
of mREM/hr. With the EXTREME duty cycle - around 10E6 : 1 - I guess that IS
megarad!!
Anyhow - we did a very simple test - exposed the Siemens EPD along side a
high rate ion chamber instrument - the HPI 1030. They both tracked
relatively well and there is no reason at all to assume that they would have
even similar rate responses - thus it is reasonable to assume that both
indeed tracked the field.
We didn't try any others.
>>This surprised me because the PD-3 is a GM tube.
Well ..... if the response is such that the probability of a count is << 1
per machine pulse, then there is little chance of dead time losses and on a
statistical basis it will track fairly well BUT when the probability of a
count starts to get within a couple of orders of magnitude of 1 per pulse -
then you'll start getting losses.
Ted de Castro
tdc@ehssun.lbl.gov
University of California Lawrence Berkeley Laboratory
Bldg B75B Rm 101
Berkeley, CA 94720
(510) 486-5256
(510) 486-7304 - FAX