[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: ED underresponse in pulsed x-ray fields
We have performed one or two qualitative tests like the one you described at
CEBAF with pulsed beam. We happen to use the SAIC PD-3 ED. In our test,
the ED responded surprisingly well, I think due to the positioning of the
dosimeters in a field consisting of mainly scattered radiation. (In the
test, the beam was 1 GeV, 60-120 Hz, 100 usec on a thick copper target)
However, I have heard from someone at LBL (I think, it's been a while) who
used the same dosimeter in a Mega-rad/hr (instantaneous) field (pulsed as
well) and they responded admirably.
This surprised me because the PD-3 is a GM tube. The saving feature of the
PD-3 may be that in its calibration regime, you do a single point
calibration at a known integrated dose - and you also expose it to a high
dose rate to apply a dead time correction (it's got an EPROM an a really
easy calibration). If I remember correctly, you can choose the dose rate
for the DT correction, so presumably, you could really blast it during
calibration to set up DT correction properly for Krad-Mrad type rates. That
doesn't explain all that's going on with the pulsed response though.
If this sounds like a plug, well, take it any way you want, I like 'em.
>
> Radsafers,
>
> We recently performed X-ray radiography at Byron using a 6 MeV
> electron accelerator (MINAC 6) with a tungsten target as the X-ray
> source. I anticipated that the electronic dosimeters we use on site
> would underrespond but was surprised to measure a 90% underresponse.
> This effect occurred with EDs placed in line-of-sight with the MINAC
> and well out of the beam. The effect was not observed in areas where
> the beam or the leakage (effective energy of 2.5 MeV) had been
> attenuated by concrete walls or floors. In these cases, the EDs were
> in lower dose rate fields and responded accurately.
>
> The experiment was not set up with research grade rigor but merely
> consisted of EDs placed side by side with ion chamber self reading
> dosimeters (SRDs) at various locations in the vicinity of the MINAC.
>
> There are several factors which probably worked in concert to produce
> the underresponse. The MINAC 6 delivers 4 usec pulses at a rate of up
> to 200 per second. This equates to a duty cycle of approximately
> 0.0008 meaning the peak dose rate is approximately 1250 times the
> effective dose rate. The EDs underrespond to high dose rates (ie.
> thousands of rem/hr). Also, the electronics may have been fooled by
> the pulsed nature of the radiation.
>
> This is just an informative message. If anyone wants further
> information or has had similar experience, I invite comments.
>
> Paul Vitalis
> ALARA Health Physicist
> Byron Nuclear Power Station
> 4450 N. German Church Rd.
> Byron, IL 61010
KW