[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re[2]: ?NVLAP category V C
John/Sandy,
Here's my two cents...
If your radiological environment includes both low and high beta
spectrum (ie. Sr/Y and Tl) then you may need to redesign your
badge.
The design requirement for distinguishing beta energies requires two
elements, one at less than about 20 mg/cm2 and another less than say 80
mg/cm2. The ratio of these two elements is used to determine the
average beta energy and the appropriate beta energy correction factor
for any given exposure. Of course, another gamma-only element is
needed and this is reasonably met by an element under the 1000 mg/cm2
filter.
Your solution of using a midway response factor may be the only option
available to you. If so I know of no magic to get blood out of a
turnip.
Mike Russell, CHP
Southern California Edison
russelmj@SCE.com
This response is based on a 4 element TLD, 2 elements of LiBO and then
2 elements using CaSO. The filtration is as follows, 7, 300, 300 and
1000. The primary basis should still be true for LiF, that the
Sr-90/Y-90 will penetrate even a 1000 mg/cm^2 filter, where the Tl-204
will only penetrate a 7 mg/cm^2 filter. Therefore, if this is also
true for a pure 3 or 4 element LiF TLD, with some appropriate
filtration, the decision matrix is easy. If there is Sr-90, it can be
seen on all the elements. If it is a Tl-204 exposure, it will only be
seen on the 7 mg/cm^2 element, the others showing basically control
background response. This is also true for Category VII, which we also
test, and again, the mixture of Cs-137 with either Tl-204 or Sr-90 is
quite evident.
What you should be doing is to define the algorithm correction factors
for specific exposures to either the Tl-204 or Sr-90. Once you
dtermine the energy of the source, the correction factor will be
accurate within a small % deviation from true exposure. There is no
reason to be high on one and low on the other. Test and let the
algorithm correct for your response of the dosimeter.
If this basis is not true, then I'm sure someone else will jump in
here, discussing the LiF response.
Sandy Perle
Supervisor Health Physics
Florida Power and Light Company
Nuclear Division
(407) 694-4219 Office
(407) 694-3706 Fax
sandy_perle@email.fpl.com
HomePage: http://www.lookup.com/homepages/54398/home.html
______________________________ Reply Separator _________________________________
Subject: ?NVLAP category V C
Author: radsafe@romulus.ehs.uiuc.edu at Internet-Mail
Date: 12/4/95 1:57 PM
I am writing my MS thesis on a LiF dosimeter design. Is anyone testing
in the beta mixture category using a standard 3 or 4 element badge? I
can just get by using a response factor midway between that for Sr/Y-90
and Tl-204, but am naturally overresponding in one case and
underresponding in the other. How are people handling this category?
My badge does not have filtration to distinguish the beta energies. The
badge is designed to handle both neutrons and photons as well, so three
elements are filtered with Cu, Cd, and plastic. A mylar window covers
the sole thin beta-detecting chip.
Please respond to zummo@gene.com
Thank you in advance for the help.
John Zummo, RSO
EH&S - MS 71
Genentech, Inc.
460 Point San Bruno Boulevard
South San Francisco CA 94080-4990
(415) 225-2959
(415) 225-5008 (fax)
zummo@gene.com