[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: RO2 ENVIRONMENTAL SPECIFICATIONS
Although I was one of the original group that drafted ANSI N323, I have for
some reason not been privileged to receive a draft of the revision.
Presumably I am now out of the loop. So, if perchance the chairman of the
revision committee sees this, how about sending me a copy so that I can
better understand the concerns of Ted Castro and others.
Ron Kathren
Washington State University
100 Sprout Road
Richland, WA 99352
>>>HOW DO YOU ADDRESS ENTRIES INTO HIGH TEMPERATURE AREAS OF THE PLANT THAT
>>>REQUIRE RADIATION MONITORING WITH THE EBERLINE RO2? THE TECHNICAL MANUAL
>>>GIVES AN UPPER BOUND OF 140 DEGREES F BASED ON THE ELECTRONICS.
>
>
>I have just recentlt been lucky enough to get a look at a draft of the next
>ANSI 323 - Instrument Cal.
>
>Although It is not yet the standard it is modeled after a European standard
>(no I don't know which one) and I'd use its comments in this regard as
>guidance.
>
>It says that when an instrument is used outside of its design range that the
>facility using it this way must test (and calibrate if necessary) the
>instrument for those conditions to document the efficacy of the procedure.
>
>Eberline CAN do temperature environmental testing - but I don't know how
>high they can go. PNL can also do testing.
>
>You may not find too many places to test that high however since that is
>beyond the ANSI N17C severe environment performance criteria.
>
>BTW - I sure am glad I don't do any surveys at YOUR place!! I am SURE the
>instrument would outlast me!!!
>
>
>
>
>