The methodology noted in the post, 

>is it representative of the EDE? I am not familiar with the
>methodology used to arrive at the equation, which is:
>EDE =(1.5*chest badge) + (0.04*collar badge).

is found on page 34 in NCRP Report 122, "Use of Personal Monitors to 
Estimate Effective Dose Equivalent and Effective Dose to Workers for 
External Exposure to Low-LET Radiation" Dec. 27. 1995.   It is an 
empirically derived formula.  Use of this formula is noted as unique to 
workers receiving radiation from medical x-rays, particularly 
fluoroscopy.  It is a modification of the work of Faulkner, K. and 
Harris, R.M. (1988) "Estimation of effective dose equivalent to staff in 
diagnostic radiology"  Phys. Med. Biol. 33, 83-91 and Webster, E.W. 
(1989) "EDE  for exposure with protective aprons" Health Phys. 56, 
568-569.

The above is the personal views of the author
Ed Tupin
RSO, Center for Devices and Radiological Health
EAT@fdadr.cdrh.fda.gov
(301) 827-1230
Fax:  (301) 594-4760