[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re[2]: Dental X-rays
Thanks to all for your comments from my message yesterday. I realized
that it was sketchy, but did not want to test system bandwidth or
reader's patience with the gory details. A brief response, in no
particular order:
We did dosimetry for panoramics, for 3 representative machines using
400-speed screen-film. Results were effective dose equivalents of
0.01 to 0.02 mSv/film. We used exposures determined empirically from
a sample of each make and model.
The high-yield criteria were designed in an attempt to reduce
unnecessary x-ray exposure in dentistry by reducing the number of
unproductive exams. It seems impossible to eliminate these entirely.
If every exam is positive, then the criteria were probably too rigid
and significant disease may have been missed.
As in many fields, dissemination and implementation of new technology
in practice lags well behind research. The long-term solution is
upgraded radiology curricula in dental schools. Radiology is not
recognized by the American Dental Association as a specialty of
dentistry, so curriculum quality varies greatly among schools. There
is a major current effort to obtain specialty recognition for dental
radiology. It was very nearly accomplished last year, and probably
will be in the next few years.
For our Monte Carlo calculations we used optimum skin exposures by
technique and beam energy, measured by Larry Crabtree at CDRH in the
mid-80s. These were determined for optimum conditions of darkroom QA,
etc., and are less than those from DENT, NEXT, etc. I don't think the
data were ever published. We used organ weighting factors from ICRP
26. The ICRP at that time was quite dogmatic that the term effective
dose equivalent could be used only with those factors. Our anatomic
input was obtained from head-to-foot CT scans of a typical adult
female cadaver. We had dose for the entire remainder. Our results
are strictly applicable to adult female patients. Other investigators
have since published results of measurements and calculations of
effective dose equivalent for dental exams. Results agree reasonably
well, given the large uncertainties of the method.
Sorry to ramble on so long--but you asked for it.
======================================================================
S. Julian Gibbs, DDS, PhD Voice 615-322-3190
Professor of Radiology FAX 615-322-3764
Vanderbilt University Medical Center
Nashville TN 37232-2675 Internet julian.gibbs@mcmail.vanderbilt.edu
======================================================================