>Consider the following hypothetical situation:
>
>A Nuclear Medicine patient was intended to receive a 29 mCi dose of NaI
>for hyperthyroidism.  The Authorized User filled out the Written Directive
>as required.  The Nuclear Medicine Technologist prepared the dose
>based on this Written Directive.  The patient identity was confirmed by
>two methods.  The dose was administered as directed.  The Nuclear
>Medicine Technologist made the Written Record as required.  It was an
>textbook case.
>
>However, unbeknownst to all, the calibration of dose calibrator for that
>particular isotope on that particular day was erroneous by more than
>20%.  (It is not important how that occurred.)  Therefore, the event is
>undeniably a misadministration.
>
>Here are my questions:
>
>1.   Would this be a violation of 10 CFR 35.32 as a failure of the QMP to
>assure that byproduct material will be administered as directed by the
>authorized user because the accuracy requirement of the dose
>calibrator is implicit in the QMP requirements?
>
>2.   Or instead, would this be a violation of 10 CFR 35.50, which
>addresses the accuracy requirement of the dose calibrator explicitly?
>
>All comments are welcome.
>