EVENT TEXT
+------------------------------------------------------------------------------+
| AGREEMENT STATE REPORT OF A MEDICAL MISADMINISTRATION AT HERMANN
HOSPITAL IN |
| HOUSTON, TEXAS (ABNORMAL
OCCURRENCE) |
|
|
| The following text is a portion of a facsimile received from the
Texas |
| Department of Health Bureau of Radiation
Control: |
|
|
| "Incident 7492 - Texas Licensee 650 - Texas is investigating.
[Abnormal |
| Occurrence] criteria
applies." |
|
|
| "INITIAL NOTIFICATION OF THERAPY MISADMINISTRATION: License
L00650" |
|
|
| "Hermann Hospital herewith notifies the Agency of a Misadministration
of |
| Radioactive Materials in a therapy
procedure." |
|
|
| "Specifically, approximately 27.3 mCi of I-131 was administered to the
wrong |
| patient at approximately 1040 a.m. [CDT] on August 4, 1999. The error
was |
| discovered at 1315 [CDT] on August 4,
1999." |
|
|
| "The patients involved were both outpatients, female Oriental's,
with |
| English as a secondary language. Patient One (for whom the therapy
was |
| intended) is approximately 55 years old, while Patient Two (who
received the |
| dose inadvertently) is approximately 64 years
old." |
|
|
| "Patient Two had completed a scheduled bone density scan and [was]
still in |
| the Nuclear Medicine area. At that time, she was mis-identified by
the |
| technologist who was to administer the I-131 dose. Patient One was
later |
| discovered in the waiting area still needing to be dosed. She
later |
| received the prescribed I-131 dose and returned
home." |
|
|
| "The nuclear medicine physicians conferred with the Patient Two's
personal |
| care physician and with Hermann Hospital's Risk Management Office.
The |
| patient was finally located at about 1600 [CDT] on August 4, 1999.
The |
| Chief Nuclear Medicine Physician, the Nuclear Pharmacy Manager and
the |
| Radiation Safety Officer proceeded to Patient Two's Home and discussed
the |
| event with her and her husband. With their consent, we initiated
the |
| administration of supersaturated Potassium Iodide (1 ml three times
per day) |
| and Furosemide (Lasix) at an initial dosage of 40 mg/day at
approximately |
| 1720 [CDT] on August 4,
1999." |
|
|
| "We are, of course, continuing our assessment of the events leading up
to |
| this misadministration and will prepare the reports required by
TAC |
|
289.252(f)(4)."
|
|
|
| (Call the NRC operations officer for state and licensee
contact |
|
information.)
|
+------------------------------------------------------------------------------+
************************************************************************
The RADSAFE Frequently Asked Questions list, archives and subscription
information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html