Many vendors supply such instruments - Eberline, Ludlum, and Bicron-NE, just to
name a few. I would recommend using alpha scintillation probes (e.g.,
Zinc-Sulfide). They provide a sufficient detection efficiency and exhibit no
beta/gamma interference.
In order to detect 10 - 20 dpm, a static count time of many seconds to a few
minutes will be required, depending mainly upon background count time,
instrument efficiency and probe size (for direct measurements corrected to 100
cm2). One cannot frisk/scan with a portable alpha probe and obtain a 10 - 20
dpm decision level. Static counts are required. Do you need to detect 10 - 20
dpm total activity? I can see the need for detecting such levels on a smear,
but that sounds low for a total measurement.
Rodney Bauman, CHP, RRPT
rodney_bauman@wssrap-host.wssrap.com
____________________Reply Separator____________________
Subject: Alpha contamination survey meters
Author: <radsafe@romulus.ehs.uiuc.edu>
Date: 03/09/2000 12:56 PM
I am interested in recommendations for a portable alpha contamination survey
instrument and probe capable of accurately detecting 10 - 20 DPM on a smear or a
direct frisk. The instrument must not have a response to beta or gamma
radiations. The instrument will be used for unconditional release of equipment
and personnel, so the response time needs to be tolerable for
personnel frisking. If anyone has a positive experience with an instrument
matching these characteristics I would appreciate your feedback.
************************************************************************