[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re: Use detector physical area, not effective area
According to NUREG 1507 and the Chapter 6 of the MARSSIM manual, you
should use the physical area of the probe, and also use another "source
efficiency" factor to account for the real life things that one
encounters - dust, moisture, rough surfaces, self-absorption, etc. This
philosophy is adopted from ISO 7503-1.
NUREG 1507 can be found at:
http://techconf.llnl.gov/radcri/1507.html
MARSSIM can be found at:
http://www.epa.gov/radiation/marssim/obtain.htm
Phil Egidi
Colorado Department of Public Health & Environment.
phil.egidi@state.co.us
>>> Steve Frantz <sfrantz@YAHOO.COM> 01/19/03 09:50PM >>>
I just had a discussion about something that I'd like to pass
along. Excuse me if this was so obvious that everyone else
already knows it.
An alpha probe has a physical area, say 76 cm^2, but a smaller
open (effective) area of, say 50 cm^2. The difference of 26 cm^2
is the part of the probe covered by the metal grid that holds
the mylar window in place.
Some think that in calculations you should use the effective
surface area of the probe, 50 cm^2, since the rest of the probe
is blocked. So if the detector had a calibration efficiency of
25% and you read 100 cpm, you would report 800 dpm/100 cm^2 (100
cpm / 0.25 * 100/50 cm^2).
But, it turns out you should be using 76 cm^2 to get the answer
of 526 dpm/cm^2, about a third less!
The reason is that the difference in areas was taken into
account during the calibration. The calibration source was also
blocked by the metal grid. The explanation is in MARSSIM
(NUREG-1575, Rev. 1) page 6-29. Here's the example I used to
convince myself....
Say my calibration source is exactly 76 cm^2 big and has a
strength of 152 dpm (4?). For simplicity, suppose I put a
detector with an 100% efficient scintillator on it. I'd read 50
cpm, because half of the alphas go down and only 50 out of 76 of
the remaining alphas make it through the grid (effective area).
Thus, I'll report an efficiency of 32.9% (50 cpm /152 dpm).
If I now measure a contaminated surface and it reads 50 cpm, the
contamination should be reported as 200 dpm/100 cm^2 (50 cpm /
0.329 * 100/76 cm^2), using the 76 cm^2; _NOT_ 304 dpm/100 cm^2
(50 cpm / 0.329 * 100/50 cm^2), using the 50 cm^2.
The reason is obvious; if we measure the calibration source with
this detector, we know we will read 50 cpm, and we know (by
NIST) that the source is 152 dpm/76 cm^2, which equals 200
dpm/100 cm^2.
If the source is smaller than the physical size, the process is
the same because the grid is over the whole surface and always
blocks 26/76 of the incoming radiation.
The difference between physical and effective area is accounted
for in the detector efficiency.
So, use the physical probe area, not the effective area to
calculate contaminatin levels. Strange, eh?
-Stephen Frantz
Reed College
sfrantz@yahoo.com
__________________________________________________
Do you Yahoo!?
Yahoo! Mail Plus - Powerful. Affordable. Sign up now.
http://mailplus.yahoo.com
************************************************************************
You are currently subscribed to the Radsafe mailing list. To
unsubscribe,
send an e-mail to Majordomo@list.vanderbilt.edu Put the text
"unsubscribe
radsafe" (no quote marks) in the body of the e-mail, with no subject
line.
You can view the Radsafe archives at http://www.vanderbilt.edu/radsafe/
************************************************************************
You are currently subscribed to the Radsafe mailing list. To unsubscribe,
send an e-mail to Majordomo@list.vanderbilt.edu Put the text "unsubscribe
radsafe" (no quote marks) in the body of the e-mail, with no subject line.
You can view the Radsafe archives at http://www.vanderbilt.edu/radsafe/