[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
RE: Puzzler: How Does Humidity Affect Counting Efficiency?
John and Radsafers,
Well, here is my second guess at the solution to this puzzler.
But first, review the facts:
The counting efficiency went up over 2 or 3 months as humidity went up.
The backgrounds stayed the same.
The correlation holds true for three TOTALLY DIFFERENT types of instruments:
liquid scintillation,
semiconductor detector (HPGe), and
gas flow proportional.
All are pulse counting instruments with a lower level discriminator or a
counting window (region of interest).
All use electricity to generate a high voltage (on the PM tube and gas
chamber) or a voltage (across the semiconductor detector).
The only connection among these three counters is the electric power supply.
My guess is that, over several months, the voltage or current supplied to
the three instruments changed.
I suggest you measure and record the voltage supplied to these instruments
over a period of time. Perhaps the voltage drops as your air conditioning
units come on line as the season gets warmer. You could also get an
Uninterruptible Power Supply (UPS) or isolation transformer that regulates
the output voltage to a constant value (e.g., 117 volts ac) even if the
input voltage drops.
Best wishes,
Wes
Wesley R. Van Pelt, PhD, CIH, CHP
Wesley R. Van Pelt Associates, Inc.
-----Original Message-----
From: owner-radsafe@list.vanderbilt.edu
[mailto:owner-radsafe@list.vanderbilt.edu] On Behalf Of John_Sukosky@DOM.COM
Sent: Wednesday, May 26, 2004 8:31 AM
To: radsafe@list.vanderbilt.edu
Subject: Puzzler: How Does Humidity Affect Counting Efficiency?
We've had problems controlling the environmental conditions in our Counting
Lab due to a degradation of the room's environmental control system.
Counting equipment calibration and setting of statistical ranges occurred
during the winter months. As the temperature changed and humidity
increased in the past few months, here in Virginia, our source checks on
some instruments have gone into the warning (> 2 Sigma) and Control (> 3
Sigma) levels. We analyzed the data with temperature, humidity and
barometric pressure. There is no correlation with temperature or
barometric pressure but almost a perfect correlation with humidity. When
humidity goes up, count rate goes up at the same rate. When humidity goes
down, count rate goes down at the same rate. Here's a list of affected
instruments and source check results:
Instrument Type Humidity/Count Rate
Tri-Carb Liquid Scintallation Counter Strong Correlation
Tennelec Gas Flow Proportional Counter Alpha - Strong Correlation
Beta - No Correlation
HPGe High Purity Germanium Strong Correlation
Does anyone have any ideas on what the mechanism could be for this effect?
We suspect that there is some common electronic component between all three
of these instruments that's affected by humidity. What we can't understand
is why the Tennelec Alpha channel is strongly affected by humidity but the
beta channel shows absolutely no correlation with humidity??? It's using
the same source (Pb-210) that's counted at the same time for checking both
alpha and beta channels.
Thanks for any ideas on this. We're really perplexed on this one!
John M. Sukosky, CHP
Dominion
Surry Power Station
(757)-365-2594 (Tieline: 8-798-2594)
************************************************************************
You are currently subscribed to the Radsafe mailing list. To
unsubscribe, send an e-mail to Majordomo@list.vanderbilt.edu Put the
text "unsubscribe radsafe" (no quote marks) in the body of the e-mail,
with no subject line. You can view the Radsafe archives at
http://www.vanderbilt.edu/radsafe/