[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re[2]: Film badge/TLD results
Right on Rick!!! The easy part is validating the processor. The
difficult task is to ensure that system processes "haven't changed."
That means establishing a Quality System with process as well as
outcome indicators. My program processed around 60,000 TLDs last year.
25% of my budget address quality assurance. When you consider that
staff makes up about 60% of budget, you can see that quality is a very
large percentage of dollars spent.
Since the quality assurance program is intended to determine process
capability is maintained within specified tolerances .. we perform:
1. Daily QC runs on all readers when processing for dose
2. Monthly blind spiking from our plants
3. Quarterly ANSI N13.11 blind testing .. all categories
4. Other inter-comparison testing
5. A multitude of statistical analysis .. such a t-testing,
probability testing, anova analysis, shewhart control charts (updated
daily), etc.
Quality must be maintained .. to ensure that your workers are provided
the most accurate and precise dose results ..taking into account the
cost/benefit analysis. There is a point where the process must be
considered stable, where further improvements are not cost effective.
You must first determine your system's process capability to see if
you can improve your system .. before you attempt to do it.
For further discussion .. please e-mail me directly.
Note to Rick ... missed seeing you at the meeting in June. maybe next
year!
Sandy Perle
Supervisor Health Physics
______________________________ Reply Separator _________________________________
Subject: Re: Film badge/TLD results
Author: radsafe@romulus.ehs.uiuc.edu at Internet-Mail
Date: 6/28/95 8:28 AM
As the performance evaluation program administrator for DOELAP, I feel
an overwhelming need to respond to Sandy's comments regarding the use
of proficiency test results and assessments to assess commercial
processors. I believe that those two items only provide a starting
point in the evaluation of a commercial processor.
(first some preaching) Your goal is to provide the best dosimetry
service you can to your customers, the people who work at your
facility. The commercial processor only processes dosimeters; they do
not estimate dose of record for you. As such, passing DOELAP or NVLAP
testing weeds out the acceptable performers from the unacceptable, but
not necessarily the good from the bad (gosh, that's hard to believe,
but it's true.) Therefore, for each processor you evaluate, you have
to look at other things like: the completness of information you
receive, the timeliness of reports and the controls put on special
processing.
After you select a processor, your job becomes even bigger. In the
DOE, for those laboratories using commercial processors, we require
three things: (1) that the laboratory can demonstrate the accuracy of
the results received from the processor; (2) that the laboratory can
identify and correct anomalous dosimetry results at every point of the
dose evaluation and (3) that the dose records for individuals contain
sufficient information to track doses back to dosimeter reports that
include processing information. For #1, we recommend a combination of
a blind audit program and a close relationship with the processor so
the user can obtain QA information (such as QC processing information,
calibration factors, ongoing test data and assessment reports.) For
#2, the DOE labs are using a 2-tier approach: (1) to identify
anomalous doses and (2) to request that the processor provide
information regarding anomalous chip response data as it occurs. #3
is pretty self-explanatory, i.e., you, as the supplier, need to
provide the documentation linking dose histories to processing
information.
Okay, after this lengthy diatribe, I acknowledge that sometimes, we
are all stuck with balancing accuracy, service and costs. So even
though, I believe that the major portion of our effort should be
placed in obtaining the best accuracy and service, the question
remains: How much do you want to spend to assure that doses below 100
mrem/year are accurate? Finally, with a nod to Sandy, you may want to
contact him regarding how to track and trend dosimeter results to
improve the performance of your program and identify ways to conduct
cost/benefit analyses.
Happy dosimetry.
Rick Cummings cumminfm@inel.gov
Usual disclaimer regarding my opinions and employer's.
"Shun advice at any price. Now, that's what I call good advice."
Piet Hein
______________________________ Reply Separator _________________________________
Subject: Film badge/TLD results
Author: RADSAFE (INELMAIL.RADSAFE) at _EMS
Date: 6/27/95 2:55 PM
Yesterday, Sandy Perle wrote:
> Anyone who is concerned with the accuracy and precision of the
> dosimetry processor they use, or are considering using, need only to
> ask them for their NVLAP Proficiency test results ANSI N13.11 (1983 or
> 1993 version). In addition I recommend that you ask them for the
> on-site assessment report that is left with the lab.
> These two items will provide enough information regarding the
> processor to make a valid assessment of their overall program.
While true in principle, one should apply a reasonable dose of salt to the
proposal "if they did well on NVLAP (or DOELAP) then your processor will give
you good results". I'm not taking issue with the validity of the NVLAP
process or with the idea of checking your processor's results, it should just
be kept in mind that a periodic controlled test does not guarantee "good"
dosimetry results.
A couple of things that come to mind in this vein:
A processor may pass NVLAP but fail DOELAP (or vice versa)...
Or they may not be able to pass in a certain category...
There is nothing forcing you to qualify in a given category, even
though you may have exposure in that category...
A processor may even use a special algorithm for reading TLDs for the
NVLAP test that's not routinely used for personnel badges...
Other problems may creep in, such as a systematic error that is introduced by a
calibration foul-up that doesn't get corrected until months after introduction.
Or their QA may simply not be good enough to give consistent results in the
"real world".
I have heard of a number of horror stories regarding dosimetry snafu's and seen
some interesting results from limited blind spikes I have performed myself.