[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
Re[2]: Radiation Detector Calibrations
NCRP 112 contains some good guidance on this topic. Here is my
opinion.
If one uses the instrument for measuring exposure rate, calibrate for
exposure rate to a similar (or approximately similar) energy source.
This should negate having to ensure that a compensated GM tube is
used. If you have a wide variety of energies, compensation becomes
very important. For ranges not able to be calibrated due to source
restrictions, use the pulser but make sure you include a "crossover
point(s).
If the instrument is used for contamination surveys only, pulsing is
practical. However, be you should verify the tube operates at high
rates with actual radiation. For halogen quenched GM tubes the gas
can "leak-out" thereby resulting in double pulsing of an extreme
over-response. The pulsing process is very good for linearity checks
of the electronics. In any case, for contamination monitoring, an
appropriate efficiency is quite important (both energy and geometry).
Just a caution, the ludlum models 2 & 3 can be ordered with a wide
variety of meter-faces. Some have cpm only, some have mR/hr and cpm,
some have mR/hr only. The faces are designed to operate with specific
probes. Almost any GM detector can be adapted to be supported by
these instruments. If this is done, user beware !!! The only scale
that has any meaning may be cpm, not informative for worker
protection...
Eric Darois, CHP
______________________________ Reply Separator _________________________________
Subject: Re: Radiation Detector Calibrations
Author: radsafe@romulus.ehs.uiuc.edu at Internet
Date: 6/26/96 12:55 PM
I would be interested in any responses you get on this subject. We have always
calibrated our instruments in mR/hr with a Cs-137 source. But no one here uses
Cs-137 and most of our usage is P-32, S-35, P-33 or I-125. For this reason,
we are considering calibration of the instrument's CPM scale using a pulser
and then determining the detection efficiency of the individual probes for
the particular isotopes in use. Does anyone out there have references
(regulatory or otherwise) that would favor one method over the other?
Chris Hogan
Health Physics Technologist
University of Arkansas for Medical Sciences
Occupational Health and Safety
cchogan@life.uams.edu
or
chogan@colan.uams.edu
All opinions are my own...
This has not been read or approved by my employer...
This has not been read or approved by my employer..
From: Eric A. Goldman
I am trying to determine the "best" way to calibrate a survey instrument...