I do not believe you would want to rely totally on the area monitor
for ensuring the beam is "off". Most of the applications I have seen
with large Co-60 irradiators require the operator to enter with a
portable dose rate indicating meter as well as verify the beam is off
by the area monitor. This is usually a conditional requirement of
either the license or radiation program manual/procedures. NRC
regulations rarely get that specific.
I believe the reasoning is that eventhough there are numerous periodic
quality checks required to ensure the area monitor is operational,
defense-in-depth is desired.
Just my thoughts.
Jim Reese
Defense Logistics Agency
jreese@ddrw.dla.mil
______________________________ Reply Separator _________________________________
Subject: Use of Portable Radiation Monitoring Equipment
Author: <radsafe@romulus.ehs.uiuc.edu > at SMTPGATE
Date: 9/19/97 12:00 PM
GROUP:
I wish to respectfully solicit the advice/expertise of RADSAFERS on the
following issue. I pose this issue to RADSAFERS because I have been out of
the NRC and medical world for some time and am not current as to the
subtleties of some present radiation safety practices in that arena. I
presently work in the DOE (industrial? -- certainly non-medical) world.
Radiation safety for industrial x-ray devices in the DOE world is
expected
to be provided in accordance with guidance contained in ANSI N43.3,
American National Standard for General Radiation Safety -- Installations
Using Non-Medical X-Ray and Sealed Gamma-Ray Source, Energies Up To 10 MeV,
January 1993. In the case of what this standard calls "Shielded
Installations," the x-ray machine operator is required to carry a
calibrated portable radiation monitoring instrument, such as an Eberline
Model RO-2, into and out of the x-ray exposure room each and every time
he/she enters/exits the x-ray room. The intent of using such a portable
x-ray detector is to provide the operator an additional means, apart from
the various kVp meters, mA meters already on the x-ray control panel, of
verifying that the x-ray beam truly of "off."
I note that 10 CFR 35.600 in the NRC world requires Co-60 teletherapy
rooms to be equipped with a room area radiation monitor. This part also
requires the radiation safety organization that has jurisdiction over such
a unit to have portable radiation detection survey instrumentation
sufficient to perform the various radiation surveys specified in 35.641.
But I note that 10 CFR 35.600 does NOT contain any explicit requirement for
the Co-60 machine operator to carry and use a portable radiation monitoring
instrument anytime he/she enters/exits the treatment room.
Therefore, my question: Is it correct in your judgment for me to
conclude
that having an area radiation monitor inside the treatment room is
sufficient to exempt the Co-60 unit operator from having to carry and use a
portable radiation monitoring instrument upon entering/exit of the room to
verify that the radiation beam is "off"?
I do not wish to seem to be splitting semantic hairs here. In my
non-medical, DOE world, the question has now come up as to whether a
properly placed, calibrated area radiation monitor properly installed
inside the x-ray exposure room can/should be regarded as the serving the
same function as an operator carrying and using a portable radiation
monitoring instrument upon each entry/exit of the exposure room.
This question may not be of wide interest to the general RADSAFE
community; therefore, I leave it to your judgment as to whether or not you
wish to post your response to the RADSAFE net or merely respond to me