[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
RE: Ion chamber versus GM
I know this comes a little late in the discussion, but I wanted to add one
caution when using the 450 P. Our program has many years of experience
using these meters, (we currently have 20 of them) and I perform the semi
annual calibrations on the instruments. While I agree that they are
durable, light, fast responding instruments capable of measuring a wide
range of intensities, we have had an occasional problem with chambers that
lose their pressure. This can result in a large under response of the
instrument. During our January calibration, I had one instrument that read
only 72 mR/hr in a 238 mR/hr field. The meter tested fine 6 months earlier
and the user had no indication that the ion chamber had depressurized. We
have experienced at least 3 of these failures over the years. More common
is the slow leaking of the chambers over time. I almost always need to
adjust calibration factors on each unit every 6 months to keep the device
reading to within 5% of true. Occasionally, some meters have had to be
sent back to Victoreen to be repressurized before they can be adjusted to
within 10% of true. (Last year I sent 6 of our 20). So while I think they
are a good instrument, I would encourage users to test them frequently
against a known radiation field, especially if you are relying on their
accuracy in a high dose rate field.
Donald E. Parry
Radiation Safety Section
MI Dept of Consumer & Industry Services
Voice: 517-241-1989; Fax: 517-241-1981
E-mail: don.parry@cis.state.mi.us
Web site: www.cis.state.mi.us/bhs/divisions/hfs/rss/
Donald E. Parry
Radiation Safety Section
MI Dept of Consumer & Industry Services
Voice: 517-241-1989; Fax: 517-241-1981
E-mail: don.parry@cis.state.mi.us
Web site: www.cis.state.mi.us/bhs/divisions/hfs/rss/