David,
I have used both Am-241 and Tc-99m. For diagnostic x-ray energies, I find
Tc-99m will give superior results, plus it is cheap. One can use a
relatively light-weight vial shield and make a handy "gamma camera" that
give a collimated but conveniently broad beam. the main concern is that it
is liquid and can be broken. In tests, I have found it to provide an
almost exact lead thickness result. For testing unknown walls, the source
to detector distance must be well controlled and consistent. I use a 1
meter distance unshielded reading, comparing to the 1 meter readings
through a barrier (I prefer to use an ion chamber, but any detector is
fine). "Walkie-Talkie" radios are great to coordinate between the person
maneuvering the source and the person taking the measurements.
My tests indicate that it will require about a Ci of Am-241 to be generally
useful, and the Am x-ray is a much higher energy than diagnostic x-ray.
For sheet lead, this means that the thickness resolution is not so good.
There are some really good Australian papers on using Am, however.
Bob
At 04:55 PM 3/13/00 -0600, you wrote:
>dear all
>
>we have a need to perform NDT of diagnostic x-ray rooms (perhaps up to 2.5
>mm lead) in remote locations i.e. need a procedure which is simple,
>versatile and very transportable. we are thinking of using a sealed source
>of Am-241 in a suitable container & a sensitive survey meter-type detector.
>
>i have 2 questions:
>
>1. is this the optimum method? what do others do?
>
>2. where can we obtain a >= 500 mCi Am-241 source?
>
>many thanks!
> ===================================
> David Thiele PhD
> Manager, Statewide Medical Physics
> Biomedical Engineering & Health Technology Services (7D)
> Royal Brisbane Hospital
> Herston Qld 4029
> Australia
> ----------------------------------------------------
> Phone: +61 7 3636 7896
> Fax: +61 7 3636 3518
> Email: thieled@health.qld.gov.au
> ===================================
>
>************************************************************************
>The RADSAFE Frequently Asked Questions list, archives and subscription
>information can be accessed at http://www.ehs.uiuc.edu/~rad/radsafe.html
Bob Wilson, Director
Radiation Safety Office
University of Kentucky
102 Animal Pathology
Lexington, KY 40546-0076
W: 606-323-6308
FAX: 606-323-4752
bmwils2@pop.uky.edu