Re: I-131 THERAPY DOSES AT NUC MED FACILITIES
With that volume of air it seems likely that your air sample may not
detect radioactivity indistinguishable from background levels. If you
are confident that your sampling arrangement is appropriate, it seems
that you can demonstrate compliance by recording the MDA of your sampling
configuration (provided the MDA is below your effluent ALARA action
level).
Side note: We note a release of 0.05 to 0.1% of the starting activity
during most radioiodination procedures. WF U of DE
On Mon, 18 Jul 1994, Abdul BENZIKRI wrote:
>
> Dear Reader:
>
> I understand that a fraction of iodine-131 therapy doses volatilizes and
> escapes the open vial. I am wondering what methods do other Nuclear
> Medicine facilities use to account for this fraction and show effluent
> concentration is within DAC.
>
> We in-line sample the exhausted air using charcoal filter. A 12,000 CFM
> outside air is introduced to the air exhaust from our Nuclear Medicine hood.
> It seems that the fraction of I-131 volatilized is diluted to BKG
> concentrations. If you sample the air using activated charcoal, have you
> detected any I-131? I thank you for your forthcoming response.
>
> Abdul Ben-Zikri, HP
> UW-Madsion